Why we use caffeine for HPLC calibration in Detector wave
ength accuracy test

Answers were Sorted based on User's Feedback



Why we use caffeine for HPLC calibration in Detector wave ength accuracy test..

Answer / eapen

Along with the above mentions
1.It can be seperated in any RP columns.
2.It is a highly polar compound.It elutes readily in a
reverse phase column. So less time is required.

3. It is highly stable in solutions for a longer time.
4. It is recommended to use Theobromine also in Detector
wavelength accuracy check.

Is This Answer Correct ?    14 Yes 2 No

Why we use caffeine for HPLC calibration in Detector wave ength accuracy test..

Answer / karunakar reddy > elico ltd

> 4 caffeiene having different wavelents it having 2
maximas
and 1 minima.
1st maxima is 205nm
2nd maxima is 273nm
minima is 245nm
and it is primary reference standard and also suggested in
pharmacopiea.

Is This Answer Correct ?    12 Yes 2 No

Why we use caffeine for HPLC calibration in Detector wave ength accuracy test..

Answer / p.dharmakkannu

Caffeine basically having two maxima, lower wave length at
about 204nm and higher wave length at about 272nm. 1 minima
at about 242nm in methanol or mixture of methanol and water
combination(50%). It is nothing costly and easily soluble.

Maximum of the UV range products will be analyzed in this nm
range only.

it is also a primary reference standard officially in
pharmacopoeia for melting point.

If it is primary standard we will have certificate for the
caffeine. The same certificate will be used for both the
purpose.

Is This Answer Correct ?    7 Yes 2 No

Why we use caffeine for HPLC calibration in Detector wave ength accuracy test..

Answer / rajesh

caffeine is a universal primary standard and highly
stability nature.

Is This Answer Correct ?    6 Yes 1 No

Why we use caffeine for HPLC calibration in Detector wave ength accuracy test..

Answer / kondam.manoher reddy

caffeiene having different wavelenths it having 2 maximas
and 1 minima.
1st maxima is 205nm
2nd maxima is 273nm
minima is 245nm
and it is primary reference standard and also suggested in
pharmacopiea.it contains non recemic CHON.It is very stable,
we can use a long time for analysis
it is not degradable.

Is This Answer Correct ?    6 Yes 1 No

Why we use caffeine for HPLC calibration in Detector wave ength accuracy test..

Answer / kishanpal singh fare labs gurg

It is a primary standard and it will detect at dual wave
length

Is This Answer Correct ?    9 Yes 5 No

Why we use caffeine for HPLC calibration in Detector wave ength accuracy test..

Answer / neha tyagi

stable in nature

Is This Answer Correct ?    5 Yes 1 No

Why we use caffeine for HPLC calibration in Detector wave ength accuracy test..

Answer / ravindra surdikar

Caffeine gives multiple wavelength responce
i.e. 205nm,245nm & 273nm
and it is stable and not degradable for a long time.

Is This Answer Correct ?    4 Yes 0 No

Why we use caffeine for HPLC calibration in Detector wave ength accuracy test..

Answer / manikyam

It will give maxima at 273nm and 205nm and also give minima at 245nm. caffeine fastely ellute in any mobilephase like water or methanol.

Is This Answer Correct ?    4 Yes 0 No

Why we use caffeine for HPLC calibration in Detector wave ength accuracy test..

Answer / ravinder reddy

Caffeine is very stable compound and it is not degraded
it will give minimum and maximum wavelengths
it is easily soluble in methanol and acetonitrile (1 : 1)
it will get any c18 columns

Is This Answer Correct ?    5 Yes 2 No

Post New Answer

More Analytical Chemistry Interview Questions

What is the pilot-plant in pharma sector.

1 Answers   Dr Reddys,


How to standardization of 1N H2SO4?

1 Answers   GNFC, Macleods,


I have compare C2H2-air and C2H2-N2O flame AAS on determination calcium. Both use same range of std to plot calibration curve. (2-6ppm) When i measure the sample with phosphate, KCl and LaCl, C2H2-N2O flame give false positive result, around 0.5ppm. When i measure the sample with phosphste, KCl and EDTA. C2H2-N2O flame also give 0.5ppm false positive. But both above mentioned sample would not give false positive when measured by C2H2-air flame. What is the reason?

0 Answers  


In gas chromatographic analysis sometimes we use headspace, why it is required instead of manually injection?

1 Answers  


How to study genotoxicity of impurities in API?

2 Answers   NCL,






Column selection in HPLC method development?

1 Answers  


how to prepare working standard in our laboratory aganist reference standard ?

4 Answers   Merit Biolabs, Unichem,


what is the difference b/w water content & moisture content

3 Answers   Dr Reddys,


How do we quantify crystaline and amarpous forms by using (NMR, XRD)spectroscopic techniques? Which any others instruments are useful for this quantification? explain

0 Answers   Graviti Pharmaceuticals, Vimta Labs,


which brands GC coloumn available in market

4 Answers  


What is the unit of Colour?

4 Answers   IPCA,


what is titrent,& titrate

2 Answers   Graviti Pharmaceuticals,


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)