Why we use caffeine for HPLC calibration in Detector wave
ength accuracy test
Answers were Sorted based on User's Feedback
Answer / eapen
Along with the above mentions
1.It can be seperated in any RP columns.
2.It is a highly polar compound.It elutes readily in a
reverse phase column. So less time is required.
3. It is highly stable in solutions for a longer time.
4. It is recommended to use Theobromine also in Detector
wavelength accuracy check.
| Is This Answer Correct ? | 14 Yes | 2 No |
Answer / karunakar reddy > elico ltd
> 4 caffeiene having different wavelents it having 2
maximas
and 1 minima.
1st maxima is 205nm
2nd maxima is 273nm
minima is 245nm
and it is primary reference standard and also suggested in
pharmacopiea.
| Is This Answer Correct ? | 12 Yes | 2 No |
Answer / p.dharmakkannu
Caffeine basically having two maxima, lower wave length at
about 204nm and higher wave length at about 272nm. 1 minima
at about 242nm in methanol or mixture of methanol and water
combination(50%). It is nothing costly and easily soluble.
Maximum of the UV range products will be analyzed in this nm
range only.
it is also a primary reference standard officially in
pharmacopoeia for melting point.
If it is primary standard we will have certificate for the
caffeine. The same certificate will be used for both the
purpose.
| Is This Answer Correct ? | 7 Yes | 2 No |
Answer / rajesh
caffeine is a universal primary standard and highly
stability nature.
| Is This Answer Correct ? | 6 Yes | 1 No |
Answer / kondam.manoher reddy
caffeiene having different wavelenths it having 2 maximas
and 1 minima.
1st maxima is 205nm
2nd maxima is 273nm
minima is 245nm
and it is primary reference standard and also suggested in
pharmacopiea.it contains non recemic CHON.It is very stable,
we can use a long time for analysis
it is not degradable.
| Is This Answer Correct ? | 6 Yes | 1 No |
Answer / kishanpal singh fare labs gurg
It is a primary standard and it will detect at dual wave
length
| Is This Answer Correct ? | 9 Yes | 5 No |
Answer / ravindra surdikar
Caffeine gives multiple wavelength responce
i.e. 205nm,245nm & 273nm
and it is stable and not degradable for a long time.
| Is This Answer Correct ? | 4 Yes | 0 No |
Answer / manikyam
It will give maxima at 273nm and 205nm and also give minima at 245nm. caffeine fastely ellute in any mobilephase like water or methanol.
| Is This Answer Correct ? | 4 Yes | 0 No |
Answer / ravinder reddy
Caffeine is very stable compound and it is not degraded
it will give minimum and maximum wavelengths
it is easily soluble in methanol and acetonitrile (1 : 1)
it will get any c18 columns
| Is This Answer Correct ? | 5 Yes | 2 No |
Do you know what Trace metal clean technique is?describe that technique?
1 Answers Mylan, Nirma, Oil and Gas Jobs, Reddy Labs,
How to study genotoxicity of impurities in API?
Who discovered X-ray
2 Answers Huntsman Corporation,
i set electrode to 7 buffer than i used it to check 4 buffer and 10 buffer if the results are not as expected than is it better to set the buffer again to 4 buffer or is there any other method available please let me know ...
what is the need of calulation of RRF of impurity in HPLC analysis?
How require to develop GC method? how to select diluent, GAS, column selection and other chromatographic conditions?
Is it necessary to validate both assay and related substances for usp official methods for products.
How to choose the which salt is suitable for mobilephase
why we inject sample TWICE in ASSAY by HPLC analysis but in RS once?
4 Answers Aurobindo, DRL, Neuland,
What is the difference between Paddle and Basket Why we are using Basket or paddle in dissolution.
7 Answers Apotex, Cipla, GSK, Intas, Zydus Cadila,
What is (Q) point dissolution?
Related substance method equivalency on control sample or spiked sample?