Answer / dr.sp goswami
+- 10% as per usp guideline mp ratio can be changedrevised
SOP must be there for reason /justification inclusion SOP
Can be modified
| Is This Answer Correct ? | 15 Yes | 1 No |
Answer / abdulkalim
± 10% is acceptable , but your internal Guidline/SOP should
define the acceptable limit/range.
But in general ± 10% is acceptable.
| Is This Answer Correct ? | 4 Yes | 0 No |
Answer / ashwani kumar mahajan (drl)
1 )+- 10% as per USP guideline mp ratio can be changed
2)SOP must be there for reason /justification inclusion SOP
Can be modified
3) Inhouse validation data shold be available for
perticular product then this change will be acceptable to
pharmaceopia .
| Is This Answer Correct ? | 4 Yes | 0 No |
Answer / arjun singh
1 )As per USP guideline the MP can be changed +- 10%, but
to get better accuracy it should not across the +- 5 % and
the same change should be written in SOP with justification
by giving all data accuired in case of changes. it should
be included in in house specification and same can be used
in routine practice
| Is This Answer Correct ? | 2 Yes | 0 No |
Answer / rajender rangu
We can adjust up to +or - 30% relative.and not mor than 10%absolute value of minor component.
Here we can do up to 60:40 or 80:20
| Is This Answer Correct ? | 0 Yes | 1 No |
Answer / vinod kumar, nektar therapeuti
As per me ±2% change in initial mobile phase composition is
fine to check the ruggedness of the method.
| Is This Answer Correct ? | 0 Yes | 6 No |
what type of question will ask in the interview of lab chemist.
0 Answers Godavari Biorefineries, Monginis,
what is the basic principle of conductivity meter
Why we are using a Kf Factor range 4 to 6
why using Potentometry What Purpose? differance between manual and potentiometry? why using Some of materials in usig potentiometry? What is use of mericuric acetate in Potentiometry
Why sometime potassium ortho phosphate mix with acetone use for sample preparation for hplc
my name is santhosh i cmpleted b.pharmacy in 2009 then i went to uk in 2010 to pg diploma business management. after i came back to india.. recently again apply to uk for pharmaceutical analysis. they refused my visa you failed to explain y u r change subject. so plz ineed answer for y change the cource
if v use HELIUM & ARGON instead of NITROGEN as a mobile phase what happen in GC SYSTEM ?
why we use polystyrene film for ftir calibration?
explain 5 components of gas chromatography ?
1 Answers College School Exams Tests, Nucon,
How we can analysed unknown sample by hplc....
why salisylic acid not using now days for disdolution calibration?
How is the Flame test performed?