How to calibrate and check thermal mass flow meter of Make thermal instrument company inc model 60?
A Thermal Instrument was flow calibrated at the factory for a certain specified medium/fluid. If their flow meters can be recalibrated in the field for fluids other than the one specified at time of manufacture, the calibration instructions are not posted on the web site.
Their web site does have a fairly long and involved calibration check for landfill gas application (methane generated in garbage dumps/landfills) under the FAQ: What is your procedure on checking the calibration of a landfill gas meter in the field? It involves on-site multiple pitot flow measurements in a traversing pattern to get a flow average, calculating m molecular weight of the gas mixture, calculating density from the ideal gas law, and then converting to velocity and finally calculating mass flow with the tube's cross sectional area. Simple, huh?
The analog output of the transmitter, the zero at 4.00mA and the span at 20.00mA, can be adjusted with a field calibration check and adjusted as necessary.
Typically, the flow rate can only be calibrated on a certified flow stand or flow cell; field calibration is not practical. Return to the factory is typical.
Carl Ellis Measure First
| Is This Answer Correct ? | 0 Yes | 0 No |
Why 4ma - 20ma signal used in instrumentation??
How to calibrate DP tramitter in flow measurement application? consider my range of flow is 0 to 2000 kg/hr ? please give calculation of , How much pressure applied on low pressure side and high pressure side please give detail of calculation .... It so important for me please reply as soon as possible
model question paper of instrumentation for Jindal GET 2010.plz reply sudhanshukumar08@yahoo.com
Isometric drawings are generated during _________________ phase of a project
what's the minimum level required of SIL (safety integrity level)to use fire and gas system?
we are having yokogawa centum cs3000 dcs in our power plant. i need some of the operator guides NOT to appear in HIS0163 AND HIS 0162 but it shoul appear in HIS0164 HOW TO DO IT?
What is electronics?
POSSITIVE PULSE IS APPLIED TO AN INVERTER INPUT.THE TIME FROM THE LEADING EDGE OF THE INPUT TO THE LEADING EDGE OF THE OUTPUT IS 10 ns. THE TIME FROM THE TRAILING EDGE OF THE INPUT TO THE TRYLING EDGE OF THE OUTPUT IS 8 ns. WHAT ARE THE VALUES OF tPLH and tPHL?
IF TX SHOWS SLOW RESPONSE, WHILE A CLOSE LOOP AND WHEN WE PUT THE CONTROLLER ON MANUAL THE PROBLEM RECTIFIED, SO WE CAN SAY THE PROBLEM IS IN
how the 2.5bar pressure able to control the 40bar steam
8 Answers IBM, PAFL, Thirumalai Chemicals,
If we interchange HP and LP tapping in drum level measurement and reverse LRV & URV correspondingly.What is the difference between the normal tapping(HP,LP) and reversed tapping(LP,HP)?
1.how can calibration temperature transmitter and flow transmitter? 2.
Civil Engineering (5086)
Mechanical Engineering (4456)
Electrical Engineering (16639)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)