why we use -ve voltages in vibration prob?
Answer / steve sabin
As a 20-year Bently Nevada employee, and as someone
responsible for answering such questions over the years
within our company as well as externally, allow me to
explain the rather interesting circumstances behind the use
of negative voltages (-24 vdc) for powering eddy current
vibration oscillator/demodulator devices.
The short answer is that when Don Bently worked on making
solid-state versions of the eddy-current measurement system
(it was actually originally designed in the 1930s by GE
engineers using vacuum tubes), he had a choice between
using N-P-N transistors or P-N-P transistors. At the time,
transistors were quite expensive, so he chose the least
expensive of the two: P-N-P (apparently, PNP transistors
they were less expensive to manufacture 50 years ago than
their NPN counterparts).
Because the circuits used PNP transistors, a negative bias
voltage was required rather than a positive bias voltage.
Don chose -18V. This was later changed to -24V to allow
more linear range from the transducer.
At that time, the industrial instrumentation community had
not yet standardized on +24 vdc, and by the time they did,
there were so many Bently Nevada eddy current vibration
sensors installed that changing to +24V rather than -24V
was not greeting with enthusiasm by users. Hence, it has
remained -24V to this day.
This was not a deliberate effort to "be different" or "non-
conformist" on the part of Bently Nevada. It was quite
literally based on which components were the least
expensive when the technology was originally introduced 50-
plus years ago.
Regards,
Steve Sabin
Marketing Communications Manager
Bently Nevada Asset Condition Monitoring
GE Energy Services
| Is This Answer Correct ? | 24 Yes | 0 No |
What is flow ,level, pressure transmitter ?and units?
how in the 2 - wire transmitter, 24v supply and the 4ma signal gets in the same path?
how to calibrate and troubleshooting of leveltroll?
Why we are using temperature and pressure as a reference when measuring flow of steam?
how can i communicate with non-national instruments device with pc via labVIEW8.2
3 Answers Google, Mico, National Instruments,
Explain the difference between differial pressure & delta pressure?
What is Alpha & Beta fctors related to Orifice Plate & its relation with flow?
I have installed a RTD whose length is 2250mm(RTD stem length), but it's upper 200 -250mm is out of the reactor, and this upper 200mm stem is surrounded by atmospheric temp. And the lower portion i.e. 2000mm is in contact with the measuring fluid, then does the final measuring temp. Gonna change as the upper 200mm RTD stem is out of reactor?
1 Answers Process Control Technologies,
Dear all, I'm manikandan I have done by DIPLOMA EEE from 2011 , I have 5 years of experience from Power plants instrumentation field , which one field is give long life & which one field is best ,I confused a lot And also I'm studying AMIE, can I choose which field from Section B examination?
Where we r using split range and ratio controller?
whats the working principle of level transmeter?
EXPLAIN SOME IMP QUESTIONS AND ANSWERS ON TOPIC ELECTRONICS,INSTRUMENTATION.........
Civil Engineering (5086)
Mechanical Engineering (4453)
Electrical Engineering (16638)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)