In a low pressure gradient pump, different mobile phase is
mixed using a valve before entering the pump head. In a
high pressure gradient pump, each different mobile phase is
delivered by an individual pump head and then mixed at the
pump outlet. The low-pressure gradient design is of lower
cost as it only use one pump head. It can also use more
types of mobile phase without significant increase of cost.
The high-pressure gradient design provides a much faster
gradient since solvent mixing point is much closer to the
column head. This is measured using delay volume. The value
can be 50-300 uL for high pressure gradient pump and can be
2 to 3 times larger for a low pressure gradient pump. A
small delay volume is important when the analysis time is
short or the flow rate is low. If the delay volume is too
large, it become impossible to obtain reproducible gradient
run since the planed composition cannot reach the column
head before a run is finished..
| Is This Answer Correct ? | 22 Yes | 4 No |
In simple definition,we can say that in low pressure gradient sample mixing with mobile phase will take place before pump and in high pressure gradient it will take place after pump.
Thanks,
B.Gangatharan,
Caplin point Ltd-unit-1
Pondicherry.
| Is This Answer Correct ? | 2 Yes | 0 No |
Answer / srinivask1985
In Low pressure gradient we are using single pump and one
more resorvoirs.
In High pressure gradient we are using one more pumps and
one more resorvoirs.
| Is This Answer Correct ? | 4 Yes | 7 No |
why we cannot analysis below 200nm in u.v?
why we do the calibration of limit of stry light in hplc & uv
1 why polyesterene film used for ftir calibration ? 2 why k2cr2o7 is used for control of abs in uv ? 3 why holmium oxide used for control of wavelength in uv
In rs method development when we are going area normalization method to dilute standard method?
void volume is depending on column or not? if yes how will you depend?
how can we decide the limit of cleaning validation? What is the criteria for it?
What is difference between silica used in TLC and HPLC column
4 Answers Cipla, Meghmani Organics,
why in hplc peaks are in positive side(it means upside) not in negative side?
what is use in calibration of conductivitimeter?
1 Answers Claris Lifesciences, SRF,
Which is the highly polar and highly non polar column in HPLC?
How did you conclude that when stressing a compond, all degradation products have eluted in chromatogram?
Why KBR is used in IR for preparation of pellete, why it shows zero dipole moment