What is the difference between low pressure gredient and high
pressure gredient in hplc instrument?
Answers were Sorted based on User's Feedback
Answer / yojana
In a low pressure gradient pump, different mobile phase is
mixed using a valve before entering the pump head. In a
high pressure gradient pump, each different mobile phase is
delivered by an individual pump head and then mixed at the
pump outlet. The low-pressure gradient design is of lower
cost as it only use one pump head. It can also use more
types of mobile phase without significant increase of cost.
The high-pressure gradient design provides a much faster
gradient since solvent mixing point is much closer to the
column head. This is measured using delay volume. The value
can be 50-300 uL for high pressure gradient pump and can be
2 to 3 times larger for a low pressure gradient pump. A
small delay volume is important when the analysis time is
short or the flow rate is low. If the delay volume is too
large, it become impossible to obtain reproducible gradient
run since the planed composition cannot reach the column
head before a run is finished..
| Is This Answer Correct ? | 22 Yes | 4 No |
In simple definition,we can say that in low pressure gradient sample mixing with mobile phase will take place before pump and in high pressure gradient it will take place after pump.
Thanks,
B.Gangatharan,
Caplin point Ltd-unit-1
Pondicherry.
| Is This Answer Correct ? | 2 Yes | 0 No |
Answer / srinivask1985
In Low pressure gradient we are using single pump and one
more resorvoirs.
In High pressure gradient we are using one more pumps and
one more resorvoirs.
| Is This Answer Correct ? | 4 Yes | 7 No |
What is the difference between kf and LOD
How will u perform degradation study in detectors other thann PDA in HPLC Methodology
How to do spike of aflatoxin in peanut to get maximum recovery
What is a difference between potency and purity?
from where i get mortar pastle for glass bottle crush? it's required for testing of it?
Hi sir if any product monograph not given known impurities (may be 5 impurities) specifications then how we require to proceed for inhouse formulation?
is a bonded water molecule in an organic material can be easlisy and accuractly determined by KF rather than LOD
how you fix the limits of impurities?
if you have given a blank water how you analyse by general analytical techniques?
Why we use caffeine for HPLC calibration in Detector wave ength accuracy test
60 Answers Arvind, Aurobindo, Cipla, Reddy Labs, Shasun Chemicals,
what is the peak purity concept?
5 Answers Graviti Pharmaceuticals, Vimta Labs,
what is difference between IR & FTIR ? how we calibarte both ? I want to know the procedure.