What is the difference between low pressure gredient and high
pressure gredient in hplc instrument?
Answer Posted / srinivask1985
In Low pressure gradient we are using single pump and one
more resorvoirs.
In High pressure gradient we are using one more pumps and
one more resorvoirs.
Is This Answer Correct ? | 4 Yes | 7 No |
Post New Answer View All Answers
before starting analytical method valodation what you checking? and how giving preference to start validation?
calibrtion procedue of LC-MS
how to qualify the impurity?
Why sodium hydroxide used for maintain pH of phosphate buffer
Related substance method equivalency on control sample or spiked sample?
What is a difference between potency and purity?
in each media we require to use SLS? how to proceed?
if you get peak in blank then what require to do?
I have compare C2H2-air and C2H2-N2O flame AAS on determination calcium. Both use same range of std to plot calibration curve. (2-6ppm) When i measure the sample with phosphate, KCl and LaCl, C2H2-N2O flame give false positive result, around 0.5ppm. When i measure the sample with phosphste, KCl and EDTA. C2H2-N2O flame also give 0.5ppm false positive. But both above mentioned sample would not give false positive when measured by C2H2-air flame. What is the reason?
what is third integration
How would you decide dissolution medium for NCE compound of class I drug
why multimedia dissolution require to do?
give clarity of linearity and range in method validation
if you have given one product then how you determine the impurity in that?
Why perchloric acid used for potentiometric titration