First take ir of that sample and judge the functional
goroups attached to that compound and also structure of
that compound.
then check solubility.
then UV scan with known concentration and find lymda max.
then selection of column.
first try with regular solvent like water, acn,
methanol.and then go to the buffers of different pH.near to
its pka value.
also different coloumn.
| Is This Answer Correct ? | 2 Yes | 0 No |
first check the structure of the compound.identify the
chromophore groups(for UV Detector),refractive index status
(for RI Detector),reactive status with other chromophore
group compounds(Derivitation method for UV.then check the
solubility status of the compound.prepare a known
concentration solution.
In HPLC select a C18 Column which can serve better for the
polar and not polar compounds.select a buffer pH Oof +/-1
unit of pka value of the compound prepare a buffer mix with
some 20% of organic solvent to chnge the elution early
(based on polarity)
inject the prepared known concentration sample solution in
to above system by observing the chromatogram procced
further by altering buffer
pH,COLUMN,FLOW,TEMPERATURE,ORGANIC RATIO Sample size etc.
but to quantify the unknown concentration we need a known
concentration of standard.
| Is This Answer Correct ? | 2 Yes | 1 No |
How to decide assay range for non pharmacopeial API analysis by HPLC? Could you give me any reference for same e.g Guidelines or Paper publications?
What is the difference between known and unknown impurities?
Why we use methanol as solvent in kf standardisation. Why not other solvent?
2 Answers Caplin Point, Epione Labs, Hetero,
what is the use of tap density apparatus? why we calculate it? what is the difference between the bulk density and tap density?
1 Answers Alchemist, Alembic, Apex,
why ph electrode is storing in KCL solution?Why the pH max is 14.0?
how you start RS method development when for new product?
Why void volume(time in seconds) 1 in hplc system suitability
why octanol used to determine the partition coefficient ?
acetic acid is weak acid & perchloric acid is strong acid.but how perchloric acid is soluble in acetic acid
differnc btw c18 and c-8 column
for which type of applications gas chromatography is used?
5 Answers Cipla, Flamingo Pharmaceuticals,
why disodium tartarate used in karl-fischer instrument ?.