how can standardized of HCl of 0.1 N
Answers were Sorted based on User's Feedback
By using 100 mg of sodium carbonate in 100ml of water,methyl
orange as an indicator.
Normality of 0.1N HCl=wt of Na2Co3*0.01/Vol.of 0.1N HCl*.0053
| Is This Answer Correct ? | 65 Yes | 10 No |
Answer / bhushan
1)Dissolve exactly 0.53 g of the sodium Carbonate in 20ml of distilled water.
2)Dilute to 100ml . Transfer 10ml of this 0.1N sodium carbonate solution to a 125 ml Erlenmeyer flask.
3)Add few drops methyl orange indicator.
4)Titrate approximately 0.1 N HCl solution into sodium carbonate.
END POINT :- Yellow to Reddish Brown
5) Boil the solution gently for 1min.and them cool at RT by runningtap water over the out side of the flask.If colour changes back to.
orange, titrate more HCl untill the first faint but permanent reddish
brown colour appears in the solution.
Calculations : -
-----------------
0.1 X 10
Normality of HCl = --------------------------
(B.R.) ml. of HCl ti
| Is This Answer Correct ? | 36 Yes | 5 No |
2 gm Anhyd. Na2Co3 + 150 ml H20 + 4-5 drops Methyl Orange
indicator & titrate with HCl & Colour change is Yellow to
Slight Pink.
Wt of Na2Co3 x 1000
Formulla = -------------------------
B.R x 53
| Is This Answer Correct ? | 37 Yes | 14 No |
Answer / abdulkalim
Standardization of 0.1 N HCl:
----------------------------
Weigh accurately 0.5 g of tromethamine which is previously
dried at 105° C for 3 hours add 50 mL of water and add 2
drops of bromocresol green indicator and titrate against
0.1 N HCl to pale yellow end point.(As per USP)
Formula = wt of tromethamine (mg)
________________________
121.14 X mL of HCL
| Is This Answer Correct ? | 18 Yes | 4 No |
Answer / shivalal banoth
For 0.1 N/M Hcl by dissolving 100 mg of sodium carbonate in 20 mL of water
M=100 mg/titer value X 0.1/ Factor(5.30 mg)
| Is This Answer Correct ? | 0 Yes | 0 No |
How we will choose pH of a mobile phase if analyte pKa value is 4
Through assay by hplc(anhydrous basis) product got 100.8% .How much the potency of that product.Moisture content is 1.7%,Sulfated ash is 0.02%,Total impurities are 0.3%
importance of internal standard and in what conditions it is used?
how would you determine the acceptance criteria for cleaning validationof product A for product B?
1 Answers Cipla, Cypher Solutions, Torrent Power,
what is the analytical procedure for COD and BoD
What is the size of capillary tube used for analysing melting point?
what is the limit of Correlation for IR test compare to the standard and test
In UV we measure absorbance , how it is happend ? explain me
why in GC calibration of linearity regression coefficient should not be less than 0.999
What is differences between the Instrument and Equipment
How we can determine water content in piperazine