why do we adjusted ph in the mobilphase
Answers were Sorted based on User's Feedback
Answer / rajkumar
To ionise or deionise the analyte during passing of the
column. At that pH only the analyte may be in ionise form
or deionise form.
| Is This Answer Correct ? | 50 Yes | 3 No |
Answer / pharmainterview
To set sample in single form whether it may ionized form or
neutral form then only we can get good peak shpe.
Regards
pharmainterviewquestions.blogspot.in
| Is This Answer Correct ? | 16 Yes | 1 No |
Answer / simhadri
It is better to adjust our mobile PH minium 1 or 2 units away from the analyte PKa,to get good peak shape and good retention.
| Is This Answer Correct ? | 8 Yes | 1 No |
Answer / vijay
The adjust ment of ph of the buffer is to purpose of a
corrert retntion time of the main peak
| Is This Answer Correct ? | 9 Yes | 8 No |
Answer / sheela p
Because at certain pH molecules are effective that's why we adjust the pH of mobile phase.
| Is This Answer Correct ? | 3 Yes | 2 No |
Answer / gopikrishna dr reddyys
at a particular ph we wil get impurity peaks and we ll take salts in the hplc water for buffer preparation that time the ph of that solution will not stable we have adjust the ph with suitable acid or base that time the conjugation bond will form then ph of buffer will stable
| Is This Answer Correct ? | 1 Yes | 0 No |
Answer / guest
The PH adujustments is depends upon the substance we r
going to use...and it will help for elute the peak
earlier..
| Is This Answer Correct ? | 0 Yes | 3 No |
What is Specificity ?
what is regration method in method validation , how we can calculate it.
what is the structure of DOWEX-50
what is the effect of injection volume calculate flow rate in perkin elmer headspace? (ccm parameter in HS)
What is Diffidence between Normal Detector and PDA Detector?
What is the calibration of uv process and preparation
key components in gas chromotagraphy.
What is the principle of Thermal conductivity detector and FID?
what is the difference between heavy metals and sulphated ash
why buffer is added in solvent system in HPLC ??
HPLC Calibration was done in Reverse Phase, but not in normal phase why explain?
why we measure PH of a solution in PH only why not POH ?
1 Answers Stellence Pharm Science,