why we wre doing linearity siudiec in hplc method validation
Answers were Sorted based on User's Feedback
Answer / k.praveen kumar macleods pharm
If the method is linear or not from certain range suppose
in assay validation the range is 50% to 150% we are
injecting minimum 6 levels and also Calculate the regration
line the correlation coefficient should be less than 0.999
Here regration line equation is y=mx+c
where y = Regration line
m = Slope
c = Intercept.
and also calculate residual sum of squares from theorical
and practical
| Is This Answer Correct ? | 13 Yes | 4 No |
Linearity is done in hplc method validation to show that the
method is linear through out the concentration range i.e.
the hplc method can give results according to the set
concentration , for instance if 1x is the area obtained for
50% solution then for 100% solution it should bee 2X i.e.
and so on. So for every hplc method lineariy is a must
| Is This Answer Correct ? | 6 Yes | 0 No |
What is the requirement for brazil in the analytical method vslidation comparr to ich?
What attracted you to analytical chemistry?
Which parameter require to do for analytical method equivalency?
In my analysis by HPLC,In sample main peak and placebo peak are eluting with shorter resolution ( almost look like doublet). how can i get max resolution ? Is changing gradient method is usefull?
What is Delay & Dead Volume
defination of wavelength & wave number
In Number of Theoretical Plates (N) = 16 (tr/w)2. where tr: retention time, and W: peak width. My Question is here in formula what is 16.Can u explain briefly? ---Veerabhadrarao.M
3 Answers Micro Labs, Richter Themis, Sun Pharma,
What is the difference between the working standard and reference standard?
what is biorelivent dissolution media?
importance of internal standard and in what conditions it is used?
What is the name of the detector used in TOC?
what are the ideal conditions for coloumn storage
3 Answers Claris Lifesciences, Srikem Laboratories,