Answer / m.sivalingeswara rao
Hplc detector calibration to be done to know the acuracy of
the emitted light wave length to the set value as well as
to check the sensitivity of the detecotr to respond with
the concentration of analyte.
To check accuracy generally caffeine solution is used
because it shows maxima and minima at particular wave
lengths(i.e maxima at 206nm,273nm and Minima at 241nm,in
this range we can check accuracy of lamp emitted light).
Sensitivity by permorming linearity test with different
concetration of caffeine solution.
Reproducibility by injecting the replicate injection of
Caffeine solution and measuring the %RSD of the responses
| Is This Answer Correct ? | 19 Yes | 0 No |
Answer / samba siva reddy.p
detector checking with caffiene soln bcz caffiene have 2 max and 1 minima wavelenths,we inject std stock soltn(caffiene soln) with diffrent wavelengths.
201 to 209= 205(maxima)
241 to 249=245(minima)
269 to277=275(maxima)
limits+ or -2
| Is This Answer Correct ? | 2 Yes | 0 No |
what is peak purity ??
How can we separate Urea & semicarbazide in single HPLC method ?
Please tell me about the pH of Polycaboxylic ether is it in the 5-6 range ever or more than 6
0 Answers Apex Encon Projects Private Limited, Cipla,
whate is difference between calibration and validation
11 Answers Alkem Labs, Cipla, Ranbaxy, Torrent Pharma, Zydus Cadila,
if the flow rate is increases then any changes in retention time in hplc
What is the difference between BD
How to determine retention time in Chromatography
if relative response factor is 1.07, to select how much value
what is lod and loq
How to make up 100ml of a solution cintaining compound A at 95µg/ml and compound B at30µg/ml for an HPLC reference standard.
Why only 1.2 million lux hours require in photostability study?
plz tall me that any pharmacopeia and guideline is par meted to use 1.68,4.01,7.00,9.00,12.45 buffer for ph meter calibration