how calibrate the hplc detector
Answers were Sorted based on User's Feedback
Answer / m.sivalingeswara rao
Hplc detector calibration to be done to know the acuracy of
the emitted light wave length to the set value as well as
to check the sensitivity of the detecotr to respond with
the concentration of analyte.
To check accuracy generally caffeine solution is used
because it shows maxima and minima at particular wave
lengths(i.e maxima at 206nm,273nm and Minima at 241nm,in
this range we can check accuracy of lamp emitted light).
Sensitivity by permorming linearity test with different
concetration of caffeine solution.
Reproducibility by injecting the replicate injection of
Caffeine solution and measuring the %RSD of the responses
Is This Answer Correct ? | 19 Yes | 0 No |
Answer / samba siva reddy.p
detector checking with caffiene soln bcz caffiene have 2 max and 1 minima wavelenths,we inject std stock soltn(caffiene soln) with diffrent wavelengths.
201 to 209= 205(maxima)
241 to 249=245(minima)
269 to277=275(maxima)
limits+ or -2
Is This Answer Correct ? | 2 Yes | 0 No |
What is the difference between the optics of UV and PDA Detectors ?
if tech transfer fails whay require to do?
How to use RF and rrf in Rs calculation( either multiply or divide)?
What is the name of the detector used in TOC?
how can u seperate polar compounds in hplc ?
1 Answers Meditab, Piramal Healthcare,
what is a drug?
What is the definition of RRF
What is the acceptance criteria for moisture balance when calibrated with sodium sulfate
why r u using hplc grade solvents
class of solvent regarding Residual solvent e.g. Acetone his class
why we are doing seal wash and needle wash?
How to select Sample concentration in RS analysis?example.