how to prepare working standard in our laboratory aganist
reference standard ?
Answers were Sorted based on User's Feedback
Answer / akhtar khan
working standardrs are prepared against reference standard
by triplicate analysis and calculated with the potency of
reference standard then mean of the same taken as the final
results
Is This Answer Correct ? | 47 Yes | 8 No |
Answer / psreddy
1.Select the raw material from the latest consignments and
identify which consignment is closest to 100% purity
2.Following tests compared with reference standard.
a.Description
b.Identification
c.Loss on drying/water content
d.Assay (Thrice)
3.Acceptance criteria :
i.The % RSD of the values obtained by both
chemists is NMT 2.0%.
ii.The % RSD of the values obtained by
individual chemist is NMT
Is This Answer Correct ? | 17 Yes | 0 No |
Answer / sidhu s
1. For the selection of material consider the assay, impurities and expiry date. Least impurities and highest assay and expiry is suitable. 2. To be perform identification, Loss on drying, RS and Assay. 3. Perform the assay in triplicate and RSD should not be more than 1.0 percentage.
Is This Answer Correct ? | 1 Yes | 2 No |
we can use expired sample for validation and analyst qualification?
What is the definition of RRF
According to ICH,India is coming under which climatic zones?
which brands HPLC COLUMN available in market
what is mean by ambient temperature?
How do you verify the primary,secondary and tertiary amines by IR-Spectroscopy.
What is the principle of hplc detectors like.. UV detector,flurosence detector, RID?
in sampling of raw material which side is pasted sampled labels
what is the difference between melting range & melting point? Which is more significant?
A ha tc is actual defination of retention time?
why we are analyzing purity on HPLC why not in gc
3 Answers Aurobindo, Reddy Labs,
How to define samples weight in moisture content