Answer / pankaj shrivastava
There are some factors and parameters by which we can
decrease the run time as....
1. by increasing flow Rate
2. by increasing lightly Column oven temperature.
3. by increasing mobile phase strength.
3. by using short length column.
| Is This Answer Correct ? | 28 Yes | 3 No |
Answer / priyanka.r
1)By changing the mobile phase composition of organic solvent(for ex:acetonitrile),then RT will shift.
2)by changing the flow rate, length of the column, temperature.
3)the next option is better choose some advanced Technic like UPLC.
| Is This Answer Correct ? | 13 Yes | 1 No |
1) By increase the mobile phase ratio of organic solvent,
2)By increasing Flow rate, 3)Go for Gradient Progrom 4)
Reduce the column Length And Particle size 5)By Changing the
pH of the mobile phase RT will shift
| Is This Answer Correct ? | 5 Yes | 0 No |
Answer / nirav
By addition of organic solvent in mobile phase, By changing
the flow rate, By changing length of the column.
| Is This Answer Correct ? | 3 Yes | 3 No |
Answer / ra
You must try for UPLC.Or otherwise you have to do hard work on the method.
| Is This Answer Correct ? | 3 Yes | 8 No |
Answer / santhu
BY ADJUSTING THE PH OF THE MOBILE PHASE WITH BUFFERS
| Is This Answer Correct ? | 1 Yes | 7 No |
how you fix the limits of impurities?
What is the reason of RT change in new column v/s old column in HPLC
What is stray light in uv what are units for stray light
Principle of single pan analytical balance
0 Answers Amneal Pharmaceuticals, Bayer, Dr Reddys,
Why do we need GC instead of HPLC? or Why do we need HPLC instead of GC?
What is the difference between void volume and delay volume in HPLC?
what are the standard methods for HPLC?
why do we use around only 60 mg of K2cr2o7 in the UV calibration -control of absorbance test?
what is biorelivent dissolution media?
differnc btw c18 and c-8 column
How can we konw that the HPLC column is endcapped? Is it possible to identify visually without COA?
In case of dissolution test, if there is variation within 6 individual units however the batch is complying as per S1 criteria, is it necessary to do investigation for the same? What should be the criteria for such variation?