Golgappa.net | Golgappa.org | BagIndia.net | BodyIndia.Com | CabIndia.net | CarsBikes.net | CarsBikes.org | CashIndia.net | ConsumerIndia.net | CookingIndia.net | DataIndia.net | DealIndia.net | EmailIndia.net | FirstTablet.com | FirstTourist.com | ForsaleIndia.net | IndiaBody.Com | IndiaCab.net | IndiaCash.net | IndiaModel.net | KidForum.net | OfficeIndia.net | PaysIndia.com | RestaurantIndia.net | RestaurantsIndia.net | SaleForum.net | SellForum.net | SoldIndia.com | StarIndia.net | TomatoCab.com | TomatoCabs.com | TownIndia.com
Interested to Buy Any Domain ? << Click Here >> for more details...


what is procedure of calibration of remote seal
transmiter, botom to top flanch 850mm, fluid liquid
density .73, transmiter fixed on a fractionation colum,
under vaccume 10mbar

Answers were Sorted based on User's Feedback



what is procedure of calibration of remote seal transmiter, botom to top flanch 850mm, fluid liqu..

Answer / yomama

Your question suggest you are using a "Differential Pressure
Transmitter" in which case since the -10 Bar vacuum is
applied to both the Hi and Low legs the -10 Bar vacuum is
neutral the Transmitter

In that case we merely multiply the distance (850 mm) x (.73
SG) = 620.5

Is This Answer Correct ?    1 Yes 0 No

what is procedure of calibration of remote seal transmiter, botom to top flanch 850mm, fluid liqu..

Answer / aftab ahmed mazari

If you installed the remote seal transmitter at flanges,
then in case empty tank /vessel, the transmitter will
showing -mA. so do the followings:
1. Make transmitter 4 mA in this case(without any pressure,
only lower low leg remote seal pressure acting)
2. Apply 850 x 0.73 =620.5 mm of water to high side of
transmitter flange and set transmitter out put 20 mA.
aftabmazari@yahoo.com

Is This Answer Correct ?    1 Yes 0 No

what is procedure of calibration of remote seal transmiter, botom to top flanch 850mm, fluid liqu..

Answer / yomama

Your question suggest you are using a "Differential Pressure
Transmitter" in which case since the -10 Bar vacuum is
applied to both the Hi and Low legs the -10 Bar vacuum is
neutral the Transmitter

In that case we merely multiply the distance (850 mm) x (.73
SG) = 620.5

Is This Answer Correct ?    0 Yes 0 No

what is procedure of calibration of remote seal transmiter, botom to top flanch 850mm, fluid liqu..

Answer / ys

close the both hp and lp iso valve(so that any pressure
should not act on tx diaphragm). now measure the input
pressure in mm by connecting hand communicator to tx.. this
is the LRV.now measure span(ie here it is 850mm).multiply
span with liquid density(ie 850*0.73).we will get 620.5.
this is your URV

Is This Answer Correct ?    0 Yes 0 No

Post New Answer

More Instrumentation Interview Questions

I am working in a process industry, in that there are several sections devided in the SCADA screens and individual sections may START independently. But sometimes i am facing problem i.e, if section 2 Stopped by the operator immediately Section 1 or Section 3 or Section 4 is getting stopped can you expain the reason behind this?

5 Answers  


What is mean by trim?

5 Answers  


Hot calibrate a level transmitter (wet & dry legs)?

1 Answers   Honeywell, IEC,


how to measure temperature from resistance of rtd?and how to measure temp from voltage of thermocouple?is there any formulas?

3 Answers  


what is the conductivity sensor output formula and what is the ph sensor output formula?

1 Answers  


Explain Gas Turbine startup and shutdown sequences

0 Answers  


what is the meaning if the control valve CV is 27,in terms of selecting procedure?

2 Answers  


what are the steps to follow force a particular input or out put from ics triplex esd system?

1 Answers  


1)how can calibrate temperature transmitter? 2)how can calibrate flow transmitter? 3)how can calibrate level transmitter?

3 Answers   ccic, Emerson, JK Pharma, NIGC, Rockwell, UltraTech,


i am working in the automation company...i want to just know how actully pid control valves are tune????...

1 Answers  


HOW TO CONVERT 0-60KG/CM2 PRESSURE TRANSMITTER TO A LEVEL TRANSMITTER? CAN WE ABLE TO DO IT? IF YES HOW TO DO IT BY USING HART?

3 Answers  


How flow can be measured with dp?

0 Answers  


Categories
  • Civil Engineering Interview Questions Civil Engineering (5086)
  • Mechanical Engineering Interview Questions Mechanical Engineering (4456)
  • Electrical Engineering Interview Questions Electrical Engineering (16639)
  • Electronics Communications Interview Questions Electronics Communications (3918)
  • Chemical Engineering Interview Questions Chemical Engineering (1095)
  • Aeronautical Engineering Interview Questions Aeronautical Engineering (239)
  • Bio Engineering Interview Questions Bio Engineering (96)
  • Metallurgy Interview Questions Metallurgy (361)
  • Industrial Engineering Interview Questions Industrial Engineering (259)
  • Instrumentation Interview Questions Instrumentation (3014)
  • Automobile Engineering Interview Questions Automobile Engineering (332)
  • Mechatronics Engineering Interview Questions Mechatronics Engineering (97)
  • Marine Engineering Interview Questions Marine Engineering (124)
  • Power Plant Engineering Interview Questions Power Plant Engineering (172)
  • Textile Engineering Interview Questions Textile Engineering (575)
  • Production Engineering Interview Questions Production Engineering (25)
  • Satellite Systems Engineering Interview Questions Satellite Systems Engineering (106)
  • Engineering AllOther Interview Questions Engineering AllOther (1379)