how to calibrate the interface leveltrol
Answer / ashish namdev, bina refinery
interface LT use in that application where two liquid mixed
in a tank. For calibration interface type LT these
instructions to be follows -
1. set upper & lower range.
2. set upper & lower density.
3. after then put weight according calibration data sheet
& note readings at every sample weights.
4. If the output have some error then give lower trim & upper
trim command to LT with weight at 0% & 100%.
5. recheck all value at different weights.
this LT work upon bouncy principal. which liquid will have
high density that will be make reference to LT.
| Is This Answer Correct ? | 3 Yes | 3 No |
Linear amplifier.
I need the intools training institute address in india. Intools subject oriented local authors books.Refer me plz
What is the diffrence between ff tx and non ff tx
what is mean by loop checking?
Square root calculation for dp flow transmitter?
What is a snubber circuit?
why we use only 4-20 ma signal as AI and not 0-20 ma?
2 Answers Honeywell, Oil and Gas Jobs,
What are the interlocks we have to check during the start up of LP/HP compressor?
How does the "as-built" function benefit plant owner/operators?
Explain about SOURCING & SINKING concepts in PLC to field sensor wiring ?
tell me about Woodward505 Function as a point of Instrument Engineer.
what is air to open and air to close control valve and difference btwn this...and how to measure milli amp by this...and what is the role of positioner and i/p converter in this..plz help me
Civil Engineering (5086)
Mechanical Engineering (4456)
Electrical Engineering (16639)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)