Answer / prabhakaran.g.s
Reverse phase and Normal phase classification is based on
the mobile phase and stationary phase (column).But we're
calibrating the HPLC for PUMP,COLUMN OVEN,AUTOINJECTOR and
DETECTOR etc.So we can perform the calibration for HPLC
with Normal phase also.But the ease availability of column
and mobile phase of Reversed phase, it is widely use for
the HPLC calibration.
| Is This Answer Correct ? | 96 Yes | 7 No |
Answer / d.sambireddy
Reverse phase and Normal phase related to the mobile phase
and column.
But we are calibrating the HPLC Instument all parameters in
Reverse phase calibration.it is equally suitable to Normal
phase analysis
No need to check to normal phase calibration
| Is This Answer Correct ? | 28 Yes | 2 No |
Answer / achyut
calibration -related with the total performance of hplc system
pump,flow , injection , accuracy & so on
in case of reverse phase, system suitability performed with well accuracy & precision because in rp volatile organic
phases are not used so the chances of getting accuacy is more
than np
| Is This Answer Correct ? | 6 Yes | 4 No |
Answer / pramod prajapati
Reverse phase and Normal Phase is related to mobile phase
and stationary phase used in HPLC,but here we are
calibrating HPLC and can be done in any phase.
| Is This Answer Correct ? | 3 Yes | 2 No |
Answer / deepak pal
WE CALIBRATE HPLC FOR REVERSE PHASE BCOZ ONLY MOBILE PHASE
IS DIFFERENT IN NORMAL PHASE AND WE USE MOSTLY REVERSE
PHASE CHROMATOGRAPHIC TECHNIQUES FOR ANALYSING SAMPLES AND
OTHER CONDITIONS REMAIN ALMOST SAME FOR NORMAL PHASE
| Is This Answer Correct ? | 0 Yes | 0 No |
Answer / b srinu
HPLC Calibration is a parameter of instrument performance in normal phase we changed column and mobile phase.
normal phase calibration not required. if we monitoring normal phase we can perform normal phase calibration.
| Is This Answer Correct ? | 0 Yes | 0 No |
Answer / sandeepkatakam
due to cost effective we are doing in RP.otherwise we can
do in both
| Is This Answer Correct ? | 4 Yes | 6 No |
Answer / zahir
In reverse phase it is easy to calibrate hplc than normal phase bec in normal phase n haxane is used it v highly evapurate so it also change in the area of the sample as time passes and n hexane is break septa of vial so we preferred reverse phase than normal phase.
| Is This Answer Correct ? | 0 Yes | 4 No |
Answer / oumspn
coz caffeine is nonpolar & we have to take rp hplc
| Is This Answer Correct ? | 0 Yes | 8 No |
What is meaning of audit trail in hplc
3 Answers DRL, Lee Pharma, Raks Pharma,
if your impurity coeluting with each other in that situation how require to set specification? is it acceptable?
why we are using 6 dosage units in dissolution, and why we are using 6 tablets in disintegation test.is there any scientific reason beyond 6?
What if impurity area in control sample coming more as compared to LOQ level of impurity ?
How to decided test concentration in GC analysis.
why we conclude the area% for the chromatographic purity?
1 Answers Amneal Pharmaceuticals, Lupin,
India is. Under which zone as per QF1.?
4 Answers Acebright, Aurobindo, Gland Pharma,
what does optical rotation and specific rotataion means,difference between them
Why we use methy Ethyl Propylparabine For HPLC Calibration
if you have given one product then how you determine the impurity in that?
viod volume and dead time
WHAT IS UNIT OF ABSORBANCE AREA IN HPLC OR UV DETECTER