Explain about Low pressure gradient and High Pressure
gradient systems in HPLC?

Answers were Sorted based on User's Feedback



Explain about Low pressure gradient and High Pressure gradient systems in HPLC?..

Answer / jayesh patel

In low pressure gradient elution mixing of solvent done
before the pumps and in high pressure gradient elution
mixing of solvents done after passing from pumps.

Is This Answer Correct ?    38 Yes 2 No

Explain about Low pressure gradient and High Pressure gradient systems in HPLC?..

Answer / raghava.mandalapu@gmail.com

Low pressure gradient means Both the solents mix before pump
High pressure gradient means Both the solents mix after pumps

Is This Answer Correct ?    27 Yes 3 No

Explain about Low pressure gradient and High Pressure gradient systems in HPLC?..

Answer / venkata narayana

Answer:
In a high pressure gradient pump, each different mobile phase is delivered by an individual pump head and then the mobile phases are mixed at the pump outlet. In a low pressure gradient pump, different mobile phases are mixed using a valve before entering the pump head.


As a result of the fact that the low-pressure gradient design uses only one pump head, it is of lower cost. It can also use more types of mobile phase without significant increase of cost.


Since solvent mixing point is much closer to the column head in the high-pressure gradient design, it provides a much faster gradient. This is measured using delay volume. The value can be 50-300 uL for high pressure gradient pump and can be 2 to 3 times larger for a low pressure gradient pump. A small delay volume is important when the analysis time is short or the flow rate is low. If the delay volume is too large, it become impossible to obtain reproducible gradient run since the planed composition cannot reach the column head before a run is finished.

Is This Answer Correct ?    20 Yes 1 No

Explain about Low pressure gradient and High Pressure gradient systems in HPLC?..

Answer / tariq

HPLC gradient elution systems can be categorized according to their configuration as high-pressure gradient elution systems or low-pressure gradient elution systems.

IN A LOW-PRESSURE GRADIENT ELUTION SYSTEM

(1) Driving the Solenoid (Mobile Phase Mixing Valve)
In a low-pressure gradient elution system, a solenoid valve switches in synchronization with the pump intake stroke to mix multiple mobile phases. The small-capacity plunger pump requires the solenoid valve to switch very frequently and is disadvantageous to making accurate concentration settings.

(2) Effects Due to Mobile Phase Compressibility
No mobile phase intake occurs in the hatched area as the pressure increases, this area gets larger and the concentration error increases.

(3) Correcting Errors Due to Solenoid Valve Response Delay
A time delay occurs from the time of signal input until the solenoid valve operation is complete.

(4) Mechanical Errors
Mechanical errors in the plunger drive mechanism (cams, etc.) affect the concentration accuracy

(5) Accuracy
The main difference between the systems is the ability of the high pressure one to generate more accurate and reproducible gradients at flow rates and within narrow ranges of mobile phase variations.
(6) Tests
In a low-pressure gradient elution system you can’t perform the test of high resolutions. It is used to perform simple protein test with low resolutions
(7) Technology
It is an old technology formerly used
(8) Outgassing
Outgassing of dissolved gases can cause more problems in low-pressure mixing so degassing of mobile phase is essential

(9) Maintenance
The low pressure gradient system required more maintenance.

(10) Ternary Or Quaternary Gradients
The low pressure system will let you run ternary gradients or quaternary if you wish, and let you quickly select up to 4 different solvents. However, you a binary high pressure system with a switch valve that will let you use or select up to four solvents.
(11) Dwell Volume
Dwell volume is the physical volume of liquid between the point of mixing and the column. On a high pressure mixing system, the mixing occurs after the pumps, and usually there is only a small dwell volume. On a low pressure mixing system, the mixing occurs before the pump, and the dwell volume includes the volume in the mixing chamber and the low pressure tubing leading to the pump, in addition to the tubing after the pump. This volume is usually much larger than with high pressure mixing, and can be 1 mL or more.
(12) Gradient Range
The optimum gradient range for the low pressure pump is between 5% to 95% organic. The binary pump shows optimum performance from 0% to 100% organic, which is advantageous for compounds eluting at very low or at very high organic percentage.
(13) Mixing Efficiency
Another advantage for the binary pump is the high mixing efficiency, which is necessary when TFA is used as a modifier.
In the end
HP mixing systems will always outperform LP mixing system under all conditions. Hence two Pumps are most suitable for high performance applications of a gradient system requiring high pressures with pulseless and accurate flow.

Is This Answer Correct ?    3 Yes 1 No

Post New Answer

More Analytical Chemistry Interview Questions

Can you please give a suitable definition and an example for understanding the words purity and potency.

6 Answers   Cipla, Malladi Drugs and Pharmaceuticals, Reddy Labs,


how the compound separate in coulunm,explain

0 Answers   Wockhardt,


In HPLC what is Single point threshold, peak purity index and minimum peak purity index? what should be consider for pure peak.

3 Answers  


why using factor 1.00639 in burette calibration of kf

1 Answers   Cipla,


From where 16 or 5.545 is derived in theoretical plates calculation formula by USP or JP/BP/DAB and what is meaning of this coefficient and impact in calculation ?

2 Answers   Nectar Lifesciences,






HI,I CLEARED BOB CLERK EXAM. MY INTERVIEW WILL BE ON 9TH OCTOBER,2010.PLEASE SEND ME INTERVIEW QUESTIONS AND ANSWERS. THANK YOU.

0 Answers   Bank of Baroda,


What is descriminatory disolution medium ?

3 Answers   Cipla, Graviti Pharmaceuticals,


Hi sir if any product monograph not given known impurities (may be 5 impurities) specifications then how we require to proceed for inhouse formulation?

0 Answers  


why we are using buffer solutionas as a mobile phase in hplc.

2 Answers  


How you start the method development of unknown compound.

1 Answers  


how require to calculate mass balance in related sub?

1 Answers  


what is mass spectra ? why it is necessary ?

1 Answers  


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)