Answer / jayesh patel
In low pressure gradient elution mixing of solvent done
before the pumps and in high pressure gradient elution
mixing of solvents done after passing from pumps.
| Is This Answer Correct ? | 38 Yes | 2 No |
Answer / raghava.mandalapu@gmail.com
Low pressure gradient means Both the solents mix before pump
High pressure gradient means Both the solents mix after pumps
| Is This Answer Correct ? | 27 Yes | 3 No |
Answer / venkata narayana
Answer:
In a high pressure gradient pump, each different mobile phase is delivered by an individual pump head and then the mobile phases are mixed at the pump outlet. In a low pressure gradient pump, different mobile phases are mixed using a valve before entering the pump head.
As a result of the fact that the low-pressure gradient design uses only one pump head, it is of lower cost. It can also use more types of mobile phase without significant increase of cost.
Since solvent mixing point is much closer to the column head in the high-pressure gradient design, it provides a much faster gradient. This is measured using delay volume. The value can be 50-300 uL for high pressure gradient pump and can be 2 to 3 times larger for a low pressure gradient pump. A small delay volume is important when the analysis time is short or the flow rate is low. If the delay volume is too large, it become impossible to obtain reproducible gradient run since the planed composition cannot reach the column head before a run is finished.
| Is This Answer Correct ? | 20 Yes | 1 No |
Answer / tariq
HPLC gradient elution systems can be categorized according to their configuration as high-pressure gradient elution systems or low-pressure gradient elution systems.
IN A LOW-PRESSURE GRADIENT ELUTION SYSTEM
(1) Driving the Solenoid (Mobile Phase Mixing Valve)
In a low-pressure gradient elution system, a solenoid valve switches in synchronization with the pump intake stroke to mix multiple mobile phases. The small-capacity plunger pump requires the solenoid valve to switch very frequently and is disadvantageous to making accurate concentration settings.
(2) Effects Due to Mobile Phase Compressibility
No mobile phase intake occurs in the hatched area as the pressure increases, this area gets larger and the concentration error increases.
(3) Correcting Errors Due to Solenoid Valve Response Delay
A time delay occurs from the time of signal input until the solenoid valve operation is complete.
(4) Mechanical Errors
Mechanical errors in the plunger drive mechanism (cams, etc.) affect the concentration accuracy
(5) Accuracy
The main difference between the systems is the ability of the high pressure one to generate more accurate and reproducible gradients at flow rates and within narrow ranges of mobile phase variations.
(6) Tests
In a low-pressure gradient elution system you can’t perform the test of high resolutions. It is used to perform simple protein test with low resolutions
(7) Technology
It is an old technology formerly used
(8) Outgassing
Outgassing of dissolved gases can cause more problems in low-pressure mixing so degassing of mobile phase is essential
(9) Maintenance
The low pressure gradient system required more maintenance.
(10) Ternary Or Quaternary Gradients
The low pressure system will let you run ternary gradients or quaternary if you wish, and let you quickly select up to 4 different solvents. However, you a binary high pressure system with a switch valve that will let you use or select up to four solvents.
(11) Dwell Volume
Dwell volume is the physical volume of liquid between the point of mixing and the column. On a high pressure mixing system, the mixing occurs after the pumps, and usually there is only a small dwell volume. On a low pressure mixing system, the mixing occurs before the pump, and the dwell volume includes the volume in the mixing chamber and the low pressure tubing leading to the pump, in addition to the tubing after the pump. This volume is usually much larger than with high pressure mixing, and can be 1 mL or more.
(12) Gradient Range
The optimum gradient range for the low pressure pump is between 5% to 95% organic. The binary pump shows optimum performance from 0% to 100% organic, which is advantageous for compounds eluting at very low or at very high organic percentage.
(13) Mixing Efficiency
Another advantage for the binary pump is the high mixing efficiency, which is necessary when TFA is used as a modifier.
In the end
HP mixing systems will always outperform LP mixing system under all conditions. Hence two Pumps are most suitable for high performance applications of a gradient system requiring high pressures with pulseless and accurate flow.
| Is This Answer Correct ? | 3 Yes | 1 No |
I want analyze my compound at 325 nm so which standard shall i use to calibrate my HPLC system? How i can assure my hplc detector is suitable or not?
who is the father of HPLC?
i want to know about headspace and interview question on headspace
1 Answers Srikem Laboratories,
2. Two grams of Benzoic acid are dissolved in 200 ml of water and extracted with 200 ml of diethyl ether. The distribution coefficient of benzoic acid is 100, and its dissociation constant is 6.5 10-5. Calculate the distribution ratio (D) of benzoic acid at pH 2, 5, and 6. 3. Calculate D at pH 2 to 10 (1 unit apart) in the above problem, and plot D versus pH.
what is different between volumetric solution and general solution
why we get negative values in uv when i placed sample which has some concetration but not in standard
How require to perform linearity as per ANVISA? What are the acceptance criteria?
what is the meaning of stress study?
How did you conclude that when stressing a compond, all degradation products have eluted in chromatogram?
in api coa contains only process impurities and in product coa contains degradation impurities?
why ur using toulene in hexane in u.v calibration?
3 Answers Biological.e.limited, Cipla,
In Number of Theoretical Plates (N) = 16 (tr/w)2. where tr: retention time, and W: peak width. My Question is here in formula what is 16.Can u explain briefly? ---Veerabhadrarao.M
3 Answers Micro Labs, Richter Themis, Sun Pharma,