injection volume accruacy test i calibration of hplc
Answers were Sorted based on User's Feedback
Answer / mr.bujji reddy kanchi
Injection volume is linear to the concetration of analyte.
the area of the peak will be increased, if the injection
volume is increased in linear way.
| Is This Answer Correct ? | 11 Yes | 2 No |
Answer / pratima bhandarkar
Connect union in to the HPLC.
Mobile phase: Methanol: Water (50:50)
Temperature : 25°C
Retention time :1min
Flow rate :0.5mL/min
Take one HPLC vial. Fill it with water. Weigh it (W1)
Make 6 injections with 50µl injection volume
Again weigh it (W2)
Mean volume = (W1-W2)/6*100 = µl
Limit = +-2% of the injection volume
| Is This Answer Correct ? | 8 Yes | 3 No |
Answer / ravi
In this calibration water factor (0.99602)to be used to
convert the weight in volume (micro liter)
Calculation should be as follows
Mean volume = (W1-W2)/6*0.99602 = µl
Limit = +-2% of the injection volume
| Is This Answer Correct ? | 5 Yes | 6 No |
give tha clarity about peak purity concept
3 Answers Aurobindo, Reliance Life Sciences,
how ir specrometer is calibrated ?
What is the difference between low pressure gredient and high pressure gredient in hplc instrument?
If we have 120 minutes run time method for HPLC and you wan to decrease the Run time so what should you do??
why we calibrate ph meter and balance daily.....but hplc and uv requires quaterly?
What if impurity area in control sample coming more as compared to LOQ level of impurity ?
WHY given much noise PDA dector then VWD Dector
what is difference of uv and pda
13 Answers Akums, MSN, MSN Pharma, Shilpa Medicare,
ash or residue in that what will be there.
what is lod and loq
Nitrogen test is used for which chromatography?
acceptance criteria for ph meter, hplc detector,pump,injector,oven when we calibrate.