Is it necessary to inject three replicate for accuracy in AMV.
Answer / s chatterjee
Your method's range is where the method is precise /accurate and repeatable.
By three injections you are generating a % RSD and this is usually given < 2.0 % for assay.
3 injection is necessary to prove repeatability of the test method at 50 % and 150 % accuracy concentration so that the range can be derived with confidence.
| Is This Answer Correct ? | 0 Yes | 0 No |
what is the importance of peak purity in HPLC and how we can calculate through manul(not software)?
In UV calibration we use Holmium perchlorate solution made in 1.4M perchloric acid The question is that 1.4 M perchloric acid we make is it in water or glacial acetic acid?
why we get only one peak in assay by hplc but in rs by hplc get more peaks
What is the correct way to digest milk powder to get the corrrect value of Fe and Zinc?
why are not used methonol in kf titrator to find out water content in ketones?
Is it required to inject 6 injections for % RSD of a system suitability in lc or gc
In BP2013, Loperamid HCl monograph. Assay by titration with 0.1N sodium hydroxide using hydrocloric acid 0.01N and reading the volume added between the 2 points of inflexion. I have a question that if the diluent solvent is ethanol is certainly consumed a amount of volume of titrant, so this volume must be eliminated on the result calculation or not apart from first point which is subtracted above.
what is separation techniques.
why we use Dichlorobenzene.nitrobezine.t-butyldi sulhate.for calibration of gc Head space
Suppose we prepared pH buffer solution in lab.and calibrate against NIST solution, and same shall going to used upto 30 days. is it possible to stable pH and apperance ? suggest ?
why salicylic acid not use for dissolution.
2 Answers Cipla, Indo Baijin Chemicals,
how you fix the limits of impurities?