What should be the calibration procedure for wavelength
accuracy in case of HPLC?

Answers were Sorted based on User's Feedback



What should be the calibration procedure for wavelength accuracy in case of HPLC?..

Answer / ankama rao.puppala

wavelength accuracy was done by the following reagents

1.Caffeine
2.Erbium Perchlorate
3.benzophenone

In case of Caffeine
we check the absorbance(Area)the following nanometers
200nm-210nm(205nm is the maximum absorbance)
270nm-280nm(273nm is the maximum absorbance)
240nm-250nm(245nm is the miminum absorbance)
we got the above parameters wavelength accuracy was
passed...


In the above Caffeine was best reagent becos of
it has Two maxima(205nm, 273nm) and one minima(245nm)
More stable, Cheaf cost more over it is recomanded by
Pharmacopeia.

Is This Answer Correct ?    17 Yes 0 No

What should be the calibration procedure for wavelength accuracy in case of HPLC?..

Answer / s.elumalai

Normally Above 240 nm Holmium Perchlorate and below 240nm
Caffeine solution are used,There are Holmium perchlorate
filters are available for waters HPLC.Inbuilt wavelength
tests are available for agilent.

Is This Answer Correct ?    2 Yes 1 No

What should be the calibration procedure for wavelength accuracy in case of HPLC?..

Answer / aslam

Gemerally 20 ppm solution of URACIL is used for wavelength
calibration at the wavelength rangning from 250nm---252nm---
--254nm-------------------------- upto 262nm to get the
linearity

Is This Answer Correct ?    2 Yes 1 No

What should be the calibration procedure for wavelength accuracy in case of HPLC?..

Answer / alpesh patel

Can you give any reference from pharmacopoeia? with chapter number...

Is This Answer Correct ?    0 Yes 0 No

Post New Answer

More Analytical Chemistry Interview Questions

What is the principle of HPLC?

3 Answers   Cipla, Lupin,


Retention time & Relative Retention time Definition & between difference? How to Calculate response factor(RF) & relative response factor(RRF)?

3 Answers   MSN Pharma, Titan,


wat is reverse phase in hplc

2 Answers  


Give me details about dissolution , with name ?

1 Answers   EPCA,


Why monograph of hplc coming upperside of drift but in IR monograph coming in drift side?

1 Answers  






Why 1.2% KCl is used for limit of stray light? and why limit of stray light is required in UV calibration?

5 Answers   Lupin, Ranbaxy,


why you are using benzoic acid for water content determination in TEA

2 Answers  


difference between IR& FTIR

4 Answers   Cadila,


Ph meter can show more than 14 ph reading? Why ph range in between 1 to 14 only?

0 Answers  


if partition coefficient value was positive means, what it indicates ?

1 Answers  


why we use 0.1% acetone in gradient calibration of HPLC?

3 Answers   Cipla, Dr Reddys, FDC,


what is mass spectra ? why it is necessary ?

1 Answers  


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)