What should be the calibration procedure for wavelength
accuracy in case of HPLC?
Answers were Sorted based on User's Feedback
Answer / ankama rao.puppala
wavelength accuracy was done by the following reagents
1.Caffeine
2.Erbium Perchlorate
3.benzophenone
In case of Caffeine
we check the absorbance(Area)the following nanometers
200nm-210nm(205nm is the maximum absorbance)
270nm-280nm(273nm is the maximum absorbance)
240nm-250nm(245nm is the miminum absorbance)
we got the above parameters wavelength accuracy was
passed...
In the above Caffeine was best reagent becos of
it has Two maxima(205nm, 273nm) and one minima(245nm)
More stable, Cheaf cost more over it is recomanded by
Pharmacopeia.
| Is This Answer Correct ? | 17 Yes | 0 No |
Answer / s.elumalai
Normally Above 240 nm Holmium Perchlorate and below 240nm
Caffeine solution are used,There are Holmium perchlorate
filters are available for waters HPLC.Inbuilt wavelength
tests are available for agilent.
| Is This Answer Correct ? | 2 Yes | 1 No |
Answer / aslam
Gemerally 20 ppm solution of URACIL is used for wavelength
calibration at the wavelength rangning from 250nm---252nm---
--254nm-------------------------- upto 262nm to get the
linearity
| Is This Answer Correct ? | 2 Yes | 1 No |
Answer / alpesh patel
Can you give any reference from pharmacopoeia? with chapter number...
| Is This Answer Correct ? | 0 Yes | 0 No |
from where i get mortar pastle for glass bottle crush? it's required for testing of it?
What are the limits for uniformity dosage content for tablets and capsules
Why are using in Hi,zero air, Ni gass are used GC
5 Answers Aurobindo, ELBS, Vimta Labs,
all types of questions asked at the cipla interview
0 Answers Aurobindo, Calyx, Cipla, Graviti Pharmaceuticals,
principle of AAS & its uses ?
which solution used for detecting factor of k.f.titroter except water
What is GF254 means in TLC plate?
why in GC calibration of linearity regression coefficient should not be less than 0.999
In rs method development when we are going area normalization method to dilute standard method?
Wt s the use of linearity in HPLC method validation
How can we determine Moisture content of HI?
why we use only vanilline , acetanilide,& caffeine For the Calibration of melting point appartus?