What should be the calibration procedure for wavelength
accuracy in case of HPLC?
Answers were Sorted based on User's Feedback
Answer / ankama rao.puppala
wavelength accuracy was done by the following reagents
1.Caffeine
2.Erbium Perchlorate
3.benzophenone
In case of Caffeine
we check the absorbance(Area)the following nanometers
200nm-210nm(205nm is the maximum absorbance)
270nm-280nm(273nm is the maximum absorbance)
240nm-250nm(245nm is the miminum absorbance)
we got the above parameters wavelength accuracy was
passed...
In the above Caffeine was best reagent becos of
it has Two maxima(205nm, 273nm) and one minima(245nm)
More stable, Cheaf cost more over it is recomanded by
Pharmacopeia.
| Is This Answer Correct ? | 17 Yes | 0 No |
Answer / s.elumalai
Normally Above 240 nm Holmium Perchlorate and below 240nm
Caffeine solution are used,There are Holmium perchlorate
filters are available for waters HPLC.Inbuilt wavelength
tests are available for agilent.
| Is This Answer Correct ? | 2 Yes | 1 No |
Answer / aslam
Gemerally 20 ppm solution of URACIL is used for wavelength
calibration at the wavelength rangning from 250nm---252nm---
--254nm-------------------------- upto 262nm to get the
linearity
| Is This Answer Correct ? | 2 Yes | 1 No |
Answer / alpesh patel
Can you give any reference from pharmacopoeia? with chapter number...
| Is This Answer Correct ? | 0 Yes | 0 No |
what is the diffrence between absorbance and absorption in the beer,s lamber formula ?
what are the different methods of analysing related substances test,what are LOD AND LOQ .HOW TO REPORT IMPURITIES WHEN LOD AND LOQ LIMITS R SPECIFIED
How do we quantify crystaline and amarpous forms by using (NMR, XRD)spectroscopic techniques? Which any others instruments are useful for this quantification? explain
0 Answers Graviti Pharmaceuticals, Vimta Labs,
What is the solvent/material used to balance spirit level in analytical balance?
5 Answers Glenmark, MMC Health Care,
how to regulate noise in hplc ?
Which std. used for standrdisation of ZnSO4?
diference b/w packed &capilary columns
In gas chromatographic analysis sometimes we use headspace, why it is required instead of manually injection?
whai is primary standard?
What are the different types of columns used in HPLC
what is the difference between potentiometric assay and hplc assay
what is difference between GC & GC-HS