What should be the calibration procedure for wavelength
accuracy in case of HPLC?
Answers were Sorted based on User's Feedback
Answer / ankama rao.puppala
wavelength accuracy was done by the following reagents
1.Caffeine
2.Erbium Perchlorate
3.benzophenone
In case of Caffeine
we check the absorbance(Area)the following nanometers
200nm-210nm(205nm is the maximum absorbance)
270nm-280nm(273nm is the maximum absorbance)
240nm-250nm(245nm is the miminum absorbance)
we got the above parameters wavelength accuracy was
passed...
In the above Caffeine was best reagent becos of
it has Two maxima(205nm, 273nm) and one minima(245nm)
More stable, Cheaf cost more over it is recomanded by
Pharmacopeia.
| Is This Answer Correct ? | 17 Yes | 0 No |
Answer / s.elumalai
Normally Above 240 nm Holmium Perchlorate and below 240nm
Caffeine solution are used,There are Holmium perchlorate
filters are available for waters HPLC.Inbuilt wavelength
tests are available for agilent.
| Is This Answer Correct ? | 2 Yes | 1 No |
Answer / aslam
Gemerally 20 ppm solution of URACIL is used for wavelength
calibration at the wavelength rangning from 250nm---252nm---
--254nm-------------------------- upto 262nm to get the
linearity
| Is This Answer Correct ? | 2 Yes | 1 No |
Answer / alpesh patel
Can you give any reference from pharmacopoeia? with chapter number...
| Is This Answer Correct ? | 0 Yes | 0 No |
what is the moisture content of Disodium hydrogen tartrate. how it can be calculted. as 16.5%
Shall i use a same hplc column for two different products.?
which situation gc hs and gc als require to use?
What is the thickness of polystyrene film?
8 Answers Pharmaffiliates Analytics and Synthetics,
How can we detect a compound in hplc which does not have any chromofore?
At what distance the sample is collected in dissolution basket?
why equlirium conatant of water have 14 value
why xterra column require to use at higher ph?
loss on drying we applied vaccum for some compounds and some compounds are at 105c why
how do start hplc method development? how much sample required for developing the method development?
3 Answers BMS, Cipla, Indoco Remedies, Sun Pharma,
what is the balance drift?
Wath is the differenc between two column HPLC wich have the samme support , lenth , pore size, etc.. But different name .for exemple : Spherisorb / lichrosorb