Wt s the use of linearity in HPLC method validation
Answers were Sorted based on User's Feedback
Answer / priyesh mistry
To establish the working range of your product/impurity of
interest with respect to the specification mentioned. Also
it is demonstreting that by doing plot for repsonse V/s
concentration linear relation must be seen. It also
justifies that my response is directly proportional to the
concentration of the analyte present.
| Is This Answer Correct ? | 40 Yes | 0 No |
we are proving the concentration of sample is directly
proportion to area response, why we have to do this means
during stability studies some time we will get above and
below the specification, here linearity is proving your
method is giving linear
| Is This Answer Correct ? | 3 Yes | 0 No |
Answer / parag parekh
TO CONFIRM THE RESPONCE of detector & instrument with the
concentration wheter the system is responding properly for
the change in concentration.
| Is This Answer Correct ? | 2 Yes | 2 No |
Answer / basha
to plot conc.vs area responce to give regression staight
line or coefficient of correlation 0.98
| Is This Answer Correct ? | 5 Yes | 10 No |
what are risk assement in the analytical qbd?
what is the limit for KF Factor ,we take factor using water we will found factor between 6 to 7 ,while on KF Reagent bottle mentioned 5 mg/ml water content.if we not use dried methanol it is effect on factor.
How to Prepare Dilute Sodium hydroxide solution?
what is rs test why we are perfoming rs test
What is the difference in OVI and Residual solvents?
3 Answers Discovery Intermediates,
effect of Ph during analytical run ?
1 Answers Lambda Therapeutic Research,
a bottle of concerntrated HCL (36.5) was labelled as 37% (w/w) and density 1.19m/l.What is hcl molarity?
As per USP u/vacuum means (how much vacuum should be applied)?
why 6 tablets use for dissolution
1 Answers Cadila, IPCA, SUN, Teva Pharmaceuticals,
What is the difference between purge septum flow and column flow in gas chromatography?
what is peak purity? how may types of methods to determine peak purity?
principle of AAS & its uses ?