why potassium dicromate use in calibration of UV VIIBLE
SPECTROCOPY ?
Answers were Sorted based on User's Feedback
Answer / sigilipalliramu@gmail.com
potassium dicromate having 570-620nm and uvhigest wavelength is600 so we can use that
Is This Answer Correct ? | 2 Yes | 0 No |
Answer / vikram
k2cr2o7 in H2So4 can give more maximum in the UV-VIS region,(at 235,257,313,350,430nm)so that we can check the whole UV-VIS region with a single run.
k2cr2o7 itself is stable and available in high purity.In dilute perchloric acid solution,it has a linear response with temperature.
Is This Answer Correct ? | 2 Yes | 0 No |
I have compare C2H2-air and C2H2-N2O flame AAS on determination calcium. Both use same range of std to plot calibration curve. (2-6ppm) When i measure the sample with phosphate, KCl and LaCl, C2H2-N2O flame give false positive result, around 0.5ppm. When i measure the sample with phosphste, KCl and EDTA. C2H2-N2O flame also give 0.5ppm false positive. But both above mentioned sample would not give false positive when measured by C2H2-air flame. What is the reason?
USP methodology, EP methodology, IP methodology, among three if possible to use one methodology for qualify working standard to use USP, EP, IP ? Please explain...
0 Answers Srini Pharmaceuticals,
What should be the precision for HPLC during technology transfer?
what is rt and rrt? how we find rrt? relation between rt and rrt/
If change in specification which parameter require to do for validation? if change in chromtographic condition then which parameter? if api change then which parameter? if change composition then which parameter? if old method not work out then whicj parameter? if additional one impurity added then which parameter of validation require to do on above each conditions? elaborate separately
What is the procedure for investigation of an OOS (Out of specification) results?
4 Answers Biocon, Dr Reddys, Reddy Labs,
What is the difference between the optics of UV and PDA Detectors ?
Why do we use KCl ,K2Cr207,Toulene & Halonium oxide for the caliberation of UV
why we use actone for Gpv test in hplc calibration
what is the relation of normality & molarity ? give me formula?
Calculate the water content in Di Sodium Tartrate (DST) based on molecular weight?
why required specificity in valdation