What is calibration procedure of gredient HPLC pump
Answers were Sorted based on User's Feedback
Answer / abhijeet gavali
You have two different methods for it
1.0 Use Propyl parabean in methanol and pure methanol
2.0 use 0.05% solution of acetone in water and water
| Is This Answer Correct ? | 14 Yes | 3 No |
Mobile phase A is 5.6mg/l of propylparaben in methanol (port
A and B)
Mobile phase B is 100% methanol (port C and D)
Run the test for 20 min by using curve 11 method
sample is Methanol
Acceptance criteria:
A/B/C is between 9.5 % to 10.5
A/B/D is between 9.5 % to 10.5
| Is This Answer Correct ? | 13 Yes | 5 No |
Answer / kuchela rao athota
in waters gradient program
TIME PUMPA% pUMPB% PUMPC% PUMPD% CURVE
0.00 50% 50% 0 0
2.00 0% 0% 50 50 11
6.00 50% 50% 0% 0 11
10.00 45% 45% 10 0 11
12.00 50% 50% 0 0% 11
14.00 45% 45% 0 10 11
16.00 50% 50% 0 0 11
18.00 50% 50% 0 0 11
| Is This Answer Correct ? | 8 Yes | 6 No |
Answer / prerna
Gradient Programme for port A and C :
00 : 100
00 : 100
25 : 75
25 : 75
50 : 50
50 : 50
75 : 25
75 : 25
100 : 00
100 : 00
Above programme also run for port B and D
An increasing stepwise piramid will appear. Integrate that
peak and also split each step of the peak.
Find out its height and area of height
By using the formula ( which I don't remeber ), mean flow
of solvent at each step.
Flow : Limit
25 (24-26)
50 (49-51)
75 (74-76)
100 (100)
If the observed value is within the acceptance limit then
that pump passes the test.
| Is This Answer Correct ? | 10 Yes | 10 No |
what is the definition of Precision.
What are the terms used during Melting Range Analysis ?
what is the cleaning validation?
How can we calibrate the burette dispencing capacity in metrorm kf titration calibration
how to set sample and standard concentration in RS method?
plz tall me that any pharmacopeia and guideline is par meted to use 1.68,4.01,7.00,9.00,12.45 buffer for ph meter calibration
how to calculate elemental impurities?
How to develop the method when we have very low amount of impurity?
Why dissolution test apparatus calibration with salycilic acid tablets was stopped by FDA&USP from the year of 2009?
why we are using polystyrene film for calibration of I.R spectroscopy?
12 Answers Abbott, Andhra Organics, Intas, Kemwel, Reddy Labs, RRR, Veritas, Zydus Cadila,
why we are adjusting the ph in buffer solution preparation
which are the sizes of capsules?