what is method development parameter in HPLC method
development
Answers were Sorted based on User's Feedback
Answer / abdulkalim
Parameters involved during initial stage of Analytical
Method development:
1. Optimization of Mobile Phase
2. Selection of Buffers
3. Selection of Ion pair reagent
4. pH of buffer/mobile phase
5. Composition of Mobile Phase
6. Selection of Column
7. Optimization of Analyte signal
8. Objectives of Seperation
8.1 Analysis time
8.2 Resolution
8.3 k factor (Capacity factor)
8.4 Peak height
8.5 Asyymetry
8.6 Theoritical Plates
9. Flow rate
| Is This Answer Correct ? | 90 Yes | 7 No |
Answer / ramana
Parameters involved during initial stage of Analytical
Method development:
1. Solubility
2. Pka Value
3. Selection of Buffers
4. pH of buffer/mobile phase
5. Composition of Mobile Phase
6. Selection of Column
7. Optimization of Analyte signal
8. Objectives of Seperation
8.1 Analysis time
8.2 Resolution
8.3 k factor (Capacity factor)
8.4 Peak height
8.5 Asyymetry
8.6 Theoritical Plates
9. Flow rate
| Is This Answer Correct ? | 54 Yes | 2 No |
Answer / yojana
1. Check solubility,if sample dissolved in water then use
reverse phase,othervise use normal phase.
2.wavelength selection
3.selection of mobile phase,buffer.
4. pH OF BUFFER &MOBILE PHASE
5. selection of column
6. check retention time
7 .Flow rate
8. check system suitability parameter like
i)theoritical plates
ii) Assymetry
iii) Resolution
iv) capacitor fator
| Is This Answer Correct ? | 31 Yes | 2 No |
Answer / vinod kumar, nektar therapeuti
1.If molecular weight less than 2000. based on their
solubility follow as below
2.If water soluble and non ionic follow reverse phase.
3.If water soluble and ionic follow reverse phase ion pair
method.
4.If organic soluble follow normal phase.
5.If molecular weight is more than 2000.Based on their
solubility follow as below.
6.If organic soluble follow gel permeation chromatography
7.If water soluble ion exchange or gel filtration.
8.Once the technique is fixed optimise column.
9.Optimise buffers
10.optimise pH of buffers
11.optimise concentration of buffers.
12.optimise type of organic solvent.
13.optimise column temperature
14.optimise gradient programme to improve resolution and to
elute out late eluting peaks.
| Is This Answer Correct ? | 19 Yes | 11 No |
Answer / sivannarayana
1.Solubility
2.Find out the UV scan For wave length
3.optimaize Buffer & pH Value Optimaize
4.Optimaize Column Temparature & optimaize sample cooler Temparature.
5.optimaize the GP For better Resolution
| Is This Answer Correct ? | 4 Yes | 0 No |
Answer / anand patel
as per my knowledge system suitability test is the method
delopment parameter
| Is This Answer Correct ? | 10 Yes | 19 No |
Suppose we prepared pH buffer solution in lab.and calibrate against NIST solution, and same shall going to used upto 30 days. is it possible to stable pH and apperance ? suggest ?
What is normality
why holmium perchloate solution in Uv calibration
Why Peak Purity is not determined in GC analysis?
What is Band broading effect explain
How to estimate potassium?
How can we calculate offset value pH measurment
In HPLC when we use Isocratic method, Gradient method in hplc?
What is chiral and why this performing in HPLC
In the gas chromatography what is used carrier gas and make up gas?
How to decided test concentration in GC analysis.
how to Calibrate FID IN GC