Answer / guest

Acceptance Criteria
---------------------

pH Meter:
-----------

Perform 2 point or 3 point calibration as per the
use/requirement.

The slope should be 95% to 105%

Where as pH measured value for known pH buffer should be
4.01 ± 0.05
7.00 ± 0.05
9.21 ± 0.05
10.00 ± 0.05
11.00 ± 0.05

HPLC Detector:
--------------

Prepare differenct concentration of caffeine solution say
10 ppm, 25 ppm, 50 ppm, 75 ppm, 100 ppm, and 125 ppm.

Inject single injection of each solution and calculate the
correlation coefficient = sholud be 0.999

HPLC Pump:
----------

a) Flow rate accuracy
b) GPV

a) Flow rate accuracy:

Calculate flow rate accuracy using water at difference flow
rate at 0.5 mL/min , 1.0 mL/min, and 3.0 mL/min at duration
of 5 minutes and calculate the volume collected with in 5
minutes duration using stop watch collected in pre-weigh
calss A volumetric flask.Apply density of water at
specified temperature.

Acceptance Criteria: For all the flow rate should be ± 5%

b) Gradient Proportianating Value (GPV)

GPV test will be perform using 5.6 mg of Propyl paraben ----
-} 1000 mL of methanol for channel A and B if we are
calibrating the Quadranaery pump and C and D channels in
methanol run the gradient composition at the specified time
to obtain the response when compared with higest response
of methanol vs propyl paraben solution should be 10 ± 0.5
i.e 9.5 - 10.5.

We can even calibrate using 0.5% Acetone in water by
evaluating the response obtained throughout the run time.

Injector:
---------

a) Injector Linearity
b) Injector precision
c) Volume accuracy

a) Injector Linearity

Inject singe injection of 25ppm caffeine solution of
injection volume 5µ, 10µ, 20µ, 50µ and 100µ. if the
injector loop is 100 µl loop where as if the injector loop
is 200 µl then go upto 200 µl for better accuracy.

Acceptance criteria: Correlation coefficient: 0.999

b) Injector Precision:

Inject 5 times of 25 ppm caffeine solution and calculate
the RSD for area.

Accetance criteria: < 1.0%

c) Volume Accuracy:

Check the injector volume accuracy by intecting the 5µl
volume for 10 times in a pre-weigh vial containing water.

Check the injector volume accuracy by intecting the 10µl
volume for 10 times in a pre-weigh vial containing water.

Acceptance criteria: 4.95 - 5.05 µl
19.5 - 20.5 µl

Column Oven:
------------

Calibrate the column oven with calibrated
thermometer/digital at 30° C and 70°C

Acceptance Criteria: 30 ± 2 °C
70 ± 2 °C

Is This Answer Correct ?    19 Yes 1 No

acceptance criteria for ph meter, hplc detector,pump,injector,oven when we calibrate...

Answer / saptale maruti

Flow rate ± 0.2% of set value
Flow rate consistancy RSD NMT1.0%
Compositional Accuaracy : Difference in Absorbance is 0.01AU
Difference in time NMT 20SEc
Delay Volume NMT 4.5 (For 4 ml Delay Volume System)
Injection Volume Accuracy is : Average volume for 18
injection is 19.6 to 20.4µl
Linearity 0.999
Wavelenght Accuracy Maxima 273± 2 nm
Minima 205 and 245± 2 nm
Oven Temperature 2 degree from set temp.

For Ph ± 0.02

Is This Answer Correct ?    7 Yes 2 No

acceptance criteria for ph meter, hplc detector,pump,injector,oven when we calibrate...

Answer / basha

For pH + OR - 0.05
HPLC-DETECTOR + OR - 1nm
Acceptance Criteria: For all the flow rate should be ± 0.02
Acceptance criteria: Correlation coefficient: 0.98
Accetance criteria for volume:1 %
Acceptance Criteria: ± 1 °C

Is This Answer Correct ?    6 Yes 7 No

Post New Answer

More Analytical Chemistry Interview Questions

What are the different zones in stability according to ICH?

1 Answers   Apotex,


Why we use caffeine for HPLC calibration in Detector wave ength accuracy test

60 Answers   Arvind, Aurobindo, Cipla, Reddy Labs, Shasun Chemicals,


How to fix the concentration for RS by HPLC

10 Answers   Intas, Reddy Labs,


from compund pka value how to select buffer and column?

2 Answers   Emcure, Zydus Cadila,


In Karl Fischer titration,What is the situation if the material tested is only soluble in water,and not soluble in the common solvents used such as methanol ?? [e.g. Iron III Hydroxide polymaltose]

1 Answers  






What are the limits for uniformity dosage content for tablets and capsules

5 Answers   Micro Labs,


If we have 120 minutes run time method for HPLC and you wan to decrease the Run time so what should you do??

6 Answers  


what is the definition of Precision.

4 Answers   Cipla,


how to correct peak when peak comes broader in shape? How to do calibration of HPLC?

2 Answers   Mankind,


Generally for Normal Phase HPLC,stationary phase is Polar.In place of polar if we take non-polar means what happened & why?

7 Answers   Macleods,


which one is better separation in high pressure liquide chromatogram orlowpressure chromato gram in hplc

0 Answers   Hetero,


Why Deuterium lamp is used as UV light Source for UV.

4 Answers  


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)