IN pH CALIBRATION,CALIBRATION IS DONE FOR pH METER OR
ELECTRODE?HOW WILL YOU JUDGE THAT THE METER TO BE
CALIBRATED WITH 4,7,9.2 BUFFERS?EXPLAIN TEMPERATURE EFFECTS
ON pH.
Answer / prabhu
Hi dear the calibration is also one of the system
suitability test,whether the system working satisfactory or
not physically so when you do the calibration for pH meter
with standard pH buffer solution (which is from NIST-it may
accurate solution for actual pH)and once if your instrument
has calibrated,you can cross verify your insrument against
those pH buffer solution (NIST) and the acceptable range is
+/- 0.02 then you can conclude for ?
once if temperature raises the pH will decrease and its not
all cases
| Is This Answer Correct ? | 18 Yes | 5 No |
Why six unit used for precision?
What is Formula of Potency calculation on Dried basis, As is basis& Anhydrous basis.
importance of internal standard and in what conditions it is used?
why do u get different drug release while doing calibration of dissolution test appratus 1& 2(paddle & basket)?
1 Answers HDFC, Medrich, Twilight Litaka Pharma,
What is Delay & Dead Volume
what is the importence and role of LCMS and LCMS/MS and their applications? which type of compounds are analysis by lcms?
what is diffrence between specificity and selecivity?
on what basis SOP's can be changed? IF mobile phase ratio is 70:30 to how much % can we change the ratio in HPLC process
what is titrent,& titrate
2 Answers Graviti Pharmaceuticals,
How we will choose pH of a mobile phase if analyte pKa value is 4
what are the basic requirement for the WHO audit?
what is the difference between system precision and method precision?
3 Answers Teva Pharmaceuticals,