how do start hplc method development?
how much sample required for developing the method
development?
Answers were Sorted based on User's Feedback
Answer / sushant
1.HPLC method development start with literature servey to
get rough idea abt development
2.then find out functionallity, molecular structure,
Polarity of particular API or molecule. from which we can
decide method development phase RPC or NPC.
3.Find out Solubility, Max UV absorbace so that we be
assure about respective diluent & wavelegh for method
development.
if we will not get UV absorbace then there is dervatisation
method.
4.Mostly many oragnic drug molecules are polar or mid polar
so we usually go for RPC.
5.satrt with mobile phase composition Use water & MEthanol
or ACN according to sollubillity in 50:50 ratio.
6.Select Column -start with C-18 column make HYpersil or
Inertsil(3 microne x 4.6 ID )10-15 cm
7. feed other parameter like wavelength,ambient Temp,
1ml/min Flow rate
8.About sample preparation for assay preparation use conc
in between 100-500 ppm as per peak response.
9.If there are more than 10 molecule in particular API go
for gardient method for better resolution & short run time
while developing any method always follow ICH parameter abt
Tailing factor, Capacity factor, Assymetry, thereotical
plates.Change MP composition ,Column.
Once you get better result then for sharpness or to save
time adjust flowrate & increase Temp.
10.If our desired peak isnot retain on water /ACN/Methanol
composition MP then go for Buffer. Phosphate buffer is
widly use buffer for its PKa value,easily Available, gives
more Hydrophillic effect, stable or non degradable for 2-3
days.
...............Before starting any method development on
HPLC Flush the system with water, check pressure, saturate
column with MP,check callibration status, lamp Hours.
Always keep data about whatever trials you have done during
method development.
Method Development is very challenging task so always be
logical before any changes in any parameter.
Mostly For HPLC method Development 10 gm sample is enough
but for overall development max 100 gm sample is sufficient.
Its a tentative quantity.
After method development method validation must be done.
| Is This Answer Correct ? | 237 Yes | 7 No |
Answer / kizar ahamed
Method development starts with solubility of the
compound,if itis soluble in water v can try with reverse
phase if is is not v should use normal phase
chromatography.
2)selection of wavelength
3) column selection
4)Gradient setting(Always develope the method by using
gradient than isocratic)
| Is This Answer Correct ? | 71 Yes | 25 No |
Answer / shantaram naik
depending upon response of peak prepare sample
concentration but ideally it is 10 ppm
| Is This Answer Correct ? | 1 Yes | 40 No |
IF WE INCREASE FLOW RATE RT DECRESE.WHY
IS THERE ANYWHERE MENTION THAT THE RE PROCESS BATCH REQUIRE TO KEPT STABILITY STYDY
why results are variation in same samples analysis in gc and hplc ?
what is karl fischer principle and reaction ,describe it
5 Answers Jubilant Organosys, Microsoft,
what is role of pH in mobile phase
3 Answers Cipla, Dhanuka Laboratories,
what is RI detector Cut-off Value
How we will diapose UV (detector) lamp?
hi i am m.sc organic chem and complete analytical lab tech dipolma now job in qc many time ask question you in m.sc organic why job in qc please help me i can not give proper ans please tell me
3 Answers Arch Pharmalabs, Natco,
what is the method of UV cabinet CALIBRTION
For standardization of volumetric solutions, acceptance limit for concentration is considered as 10% of nominal concentration and %RSD as NMT 0.2% (for appropriate no. of replicates). But , if %RSD has not met, what should be further action?
which lights using during force degradation?
How we get DST moisture 15.4-16.0%. what is the calculation or method behind it.