gc Calibration % RSD failed what is the possible reasons?
Answers were Sorted based on User's Feedback
Answer / hiren
First is not column proper installations
Secound is not proper work syrring
Thred is injector septa so much punching
Is This Answer Correct ? | 21 Yes | 0 No |
Answer / yogesh patel
flow rate of gas not proper its cose fluctuation in assay
Is This Answer Correct ? | 5 Yes | 0 No |
Answer / alasyam ramesh
1) Improper stabilisation of column and system.
2) injector septa bleeding
3) partial blockage of detector jet.
4) syring struck due to solvents nature( solvent viscosity).
5) plunger moving and taking improper injection volume.
6) some time may effect room temperature also
Is This Answer Correct ? | 5 Yes | 0 No |
If %RSD is failing in GC calibration means we have to check & replace following things,
1.Injector septa
2.glass wools in liner
3.syringe
Thanks,
B.Gangatharan.
Is This Answer Correct ? | 4 Yes | 0 No |
In Related Substances Test how we can reside the limit of impurity ?
What is primary & Secondary Standard ? Give Example
How can we calculates the solubility of the drug in different medias by injecting in to hplc????(solubility studies)
2 Answers Super Pharmaceuticals,
Difference between the quantitative analysis and qualitative analysis?
What is Endcapped Hplc column
is there any material other than caffene to calibrate hplc detector
Why use the Omeprazole as a reference standard to estimate the assay of esomeprazole by HPLC method
In kf constant term 16 what it is explain ?
What is the importance of rrf in rs by HPLC?
What is linearity in HPLC and how will you test?
What is the formula for relative diffrence for standard solution in solution stability in validation?
What is use of LOD and LOQ in validtion and give range and in calculation why we use 3.3* LOD here why we use 3.3
2 Answers Dr Reddys, IPCA, Reddy Labs,