Answer / hiren
First is not column proper installations
Secound is not proper work syrring
Thred is injector septa so much punching
| Is This Answer Correct ? | 22 Yes | 0 No |
Answer / yogesh patel
flow rate of gas not proper its cose fluctuation in assay
| Is This Answer Correct ? | 5 Yes | 0 No |
Answer / alasyam ramesh
1) Improper stabilisation of column and system.
2) injector septa bleeding
3) partial blockage of detector jet.
4) syring struck due to solvents nature( solvent viscosity).
5) plunger moving and taking improper injection volume.
6) some time may effect room temperature also
| Is This Answer Correct ? | 5 Yes | 0 No |
If %RSD is failing in GC calibration means we have to check & replace following things,
1.Injector septa
2.glass wools in liner
3.syringe
Thanks,
B.Gangatharan.
| Is This Answer Correct ? | 4 Yes | 0 No |
Why used 0.02% v/v solution of toluene in hexane for calibration of UV.
what is biorelivent dissolution media?
How to do regeneration of Metacarb Pb plus column?
why we use TLC?
How does HPLC Calculate Void volume?
What is the difference between Discriminating media and DPDM(Dissolution Profile with Different Media)
How do we fix the sample concentaryion in hplc method development. What is the basis?
Which is the highly polar and highly non polar column in HPLC?
how can you decide API molucule acedic or basic compound by test?
why we are taking 89ml concentrated HCL for preparing 1N HCL
How we fix the validity period of a Normal solution and re-standardization due date
How can you calculate manually "Slope" in pH calibration?