Condensation Test
Use this test whenever injector or carrier gas contamination
problems are suspected (e.g., ghost peaks or erratic baselines).
* Leave the GC at 40-50�C for 8 or more hours.
* Run a blank analysis (i.e., start the GC, but with no
injection) using the normal temperature conditions and
instrument settings.
* Collect the chromatogram for this blank run.
* Immediately repeat the blank run as soon as the first
one is completed. Do not allow more than 5 minutes to elapse
before starting the second blank run.
* Collect the chromatogram for the second blank run and
compare it to the first chromatogram.
* If the FIRST chromatogram contains a substantially
larger amount of peaks and baseline instability, then that
is an indication that there is contamination upstream of the
capillary column (ie. contaminated inlet, dirty carrier gas,
etc.).
* If BOTH chromatograms contain few peaks or very little
baseline drift, it can be assumed that the carrier gas
and/or inlet are relatively clean.
* If BOTH chromatograms contain a significant amount of
noise and/or baseline drift, then that usually is an
indication that the detector or detector gases are contaminated.
Is This Answer Correct ? | 8 Yes | 1 No |
Answer / guest
retention time per puss we r using mobile phase ph
Is This Answer Correct ? | 0 Yes | 0 No |
why we get only one peak in assay by hplc but in rs by hplc get more peaks
WHAT IS UNIT OF ABSORBANCE AREA IN HPLC OR UV DETECTER
iam using ph instrument lab india make.3ponit and 5point.in calibration time 3point slope will come 98.89and 96.87.that limit is 85 to 105.what basis slope will come.any calucalation between them.5point wiicome slope 99.89?why 3p0int came 2 values
what is weight veriation in pharma industry ? plz explain me
How to decided test concentration in GC analysis.
what is mean by least count balance?
Why will a solution of aspirin without acid or basic hydrolysis give yellow colour with a solution of Ferric chloride?
why buffer is added in solvent system in HPLC ??
in which situation require to take incident in validation?
WHAT IS THE PURPOSE WE USE HCL, CHOPPER, BURNER, MONOCHROMATOR in AAS? PLEASE EXPLAIN BRIEFLY
why glutent are detected in the rice cereal baby food product even manufacturer claimed that they are using rice and milk only?we have using ELISA to do the test,and rice supposed not containing any glutent,rite?We already repeat the test so many times and it still detected.just wondering where the glutent came from?
Calculate the water content in Di Sodium Tartrate (DST) based on molecular weight?