Answer / sri krishna

Condensation Test
Use this test whenever injector or carrier gas contamination
problems are suspected (e.g., ghost peaks or erratic baselines).

* Leave the GC at 40-50�C for 8 or more hours.
* Run a blank analysis (i.e., start the GC, but with no
injection) using the normal temperature conditions and
instrument settings.
* Collect the chromatogram for this blank run.
* Immediately repeat the blank run as soon as the first
one is completed. Do not allow more than 5 minutes to elapse
before starting the second blank run.
* Collect the chromatogram for the second blank run and
compare it to the first chromatogram.
* If the FIRST chromatogram contains a substantially
larger amount of peaks and baseline instability, then that
is an indication that there is contamination upstream of the
capillary column (ie. contaminated inlet, dirty carrier gas,
etc.).
* If BOTH chromatograms contain few peaks or very little
baseline drift, it can be assumed that the carrier gas
and/or inlet are relatively clean.
* If BOTH chromatograms contain a significant amount of
noise and/or baseline drift, then that usually is an
indication that the detector or detector gases are contaminated.

Is This Answer Correct ?    8 Yes 1 No

In GC, What does 'Column condensation test' mean by?..

Answer / guest

retention time per puss we r using mobile phase ph

Is This Answer Correct ?    0 Yes 0 No

Post New Answer

More Analytical Chemistry Interview Questions

what is karl fischer principle and reaction ,describe it

5 Answers   Jubilant Organosys, Microsoft,


why PH range 0-14 explain

5 Answers   DRL, Reddy Labs, Saint Gobain, Viva,


What is the procedure for UV-spectrophotometer calibration?

6 Answers  


How you develop a method in HPLC?

0 Answers   CU, Perrigo, Wockhardt,


on what basis SOP's can be changed? IF mobile phase ratio is 70:30 to how much % can we change the ratio in HPLC process

7 Answers   Cipla, Ranbaxy,






signifcance of detector lenearity in calibration f hplc

3 Answers  


why kf factor coming 0.5

5 Answers   Alkem Labs,


what is the moisture content of Disodium hydrogen tartrate. how it can be calculted. as 16.5%

1 Answers  


In the HPLC Calibration done wavelength accuracy done between 200nm-280nm .but not done remaining 300-400nm not done ?

0 Answers  


if v use HELIUM & ARGON instead of NITROGEN as a mobile phase what happen in GC SYSTEM ?

5 Answers   Cipla, Reddy Labs,


Why forced degradation studies are performed in hplc method validation?

6 Answers   Apex,


Why will a solution of aspirin without acid or basic hydrolysis give yellow colour with a solution of Ferric chloride?

1 Answers  


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)