Answer / sri krishna
Condensation Test
Use this test whenever injector or carrier gas contamination
problems are suspected (e.g., ghost peaks or erratic baselines).
* Leave the GC at 40-50�C for 8 or more hours.
* Run a blank analysis (i.e., start the GC, but with no
injection) using the normal temperature conditions and
instrument settings.
* Collect the chromatogram for this blank run.
* Immediately repeat the blank run as soon as the first
one is completed. Do not allow more than 5 minutes to elapse
before starting the second blank run.
* Collect the chromatogram for the second blank run and
compare it to the first chromatogram.
* If the FIRST chromatogram contains a substantially
larger amount of peaks and baseline instability, then that
is an indication that there is contamination upstream of the
capillary column (ie. contaminated inlet, dirty carrier gas,
etc.).
* If BOTH chromatograms contain few peaks or very little
baseline drift, it can be assumed that the carrier gas
and/or inlet are relatively clean.
* If BOTH chromatograms contain a significant amount of
noise and/or baseline drift, then that usually is an
indication that the detector or detector gases are contaminated.
| Is This Answer Correct ? | 8 Yes | 1 No |
Answer / guest
retention time per puss we r using mobile phase ph
| Is This Answer Correct ? | 0 Yes | 0 No |
what is the difference between capillary and glass column in GC
How to calculate the signal to noise ratio.
how we can identify the impurity is coming below loq at transfering site?
what r the reasons to get peak telling?
IS THERE ANY EQUIPMENT TO CHECK AND CALCULATE THE POLARITY OF A LIQUID?
How to calculate LOD & LOQ in HPLC Validation Method
14 Answers Dr Reddys, GSK GlaxoSmithKline, Reddy Labs, Reliance, UIS,
How to calibrate glassware against std calibrated glassware?
how calibrate the hplc detector
2 Answers Cipla, Shraddha Analytical Services,
difference between record & document ?
can some other acids like nitric acid or HCl can be used in the calibration of UV.Vis. spectrophotometer????
Why we are use 0.005M Sulphuric acid, in control of absorbance performance test UV Spectrophotometer calibration
What is the use of tlc and hplc? And when and where use?