How to choose concentration of LOQ solutions for RS analysis? Example.
Answer / sachin patil
Its depend on your test concentration and the specification limit, means if specify limit is 1.0% in quality and in control have 0.5% then your LOQ concentration must be below the 50% of your specify limit of of control and % RSD of recovery at this LOQ level must be performed and should be pass.
| Is This Answer Correct ? | 0 Yes | 0 No |
why ph is between 0 to 14
6 Answers Apotex, Reliance, Zydus Cadila,
why salisylic acid not using now days for disdolution calibration?
mixed 1.0 ml of methanol and 1.0 ml of toluene and injected in gas chromatography.will you prove that both solvents are in the ratio 1:1(not by area normalization)
what are risk assement in the analytical qbd?
What is the diffrence in japan mkt requirement in analytical method validation over US?
what is defferance between dipolemoment and polarity
why we are taking 89ml concentrated HCL for preparing 1N HCL
mode of absorption in alimentary canal?
solvents used in NMR
What is mean by peak tailing in case of GC?
My question about gas chromatography sulfur chemiluminsecence detector. I test unknown sample gas by GC-SCD (calibrated ) and the result of *H2S is 279 PPM , *but when I test the same sample with the GC-TCD (calibrated ) the value of *H2S is 2500 PPM . I'd like to inform you that both GCs are calibrated and have very good operation conditions with stable parameters . the question is if the sample gas with higher H2S over detection limits of SCD detector (1000 ppm). why the result it 279 ppm Best regards
if rsd failed then what require to do?