we calibrate HPLC once in 6 months while iam doing
calibration i came to know that calibration was failed.
then how can i justify that the samples analyzed on the
system during this 6 months was wright? how can i forward in
to the next step for those batches?
Answers were Sorted based on User's Feedback
Answer / ravi narwade
Whenever we are doing any analysis we are performing the
system suitability test for it, and if the "SST" is passing
no one will raise the question on analysis of samples on
the instrument which fails in calibration.
Is This Answer Correct ? | 31 Yes | 8 No |
Answer / nirchana
if u hv doubt ..thn u can check previous sample tht result u known..tht means any std sample thn match with recent sample
Is This Answer Correct ? | 2 Yes | 2 No |
Answer / k l reddy
Need not to analyze all previous batches but you should
analyze 6 previous batches(recent batches of before
calibration)on different system(which is qualified) and
compare the results with old results. Fallow the same
prociser until the resukts matched with old results.
Is This Answer Correct ? | 2 Yes | 2 No |
Answer / sasikala
Need not to analyze all previous batches but you should
have internal sop which describes how many batches
(immediate previous batches) to be analyzed and how to be
compared. But this anwser not complete one for you.
Is This Answer Correct ? | 1 Yes | 4 No |
Answer / pankaj punjabi
After 6 month calibration is failed,So you have to
reanalyse the previous batches which is analyse for last 6
month.and comapre the results.then only it is justified.
Is This Answer Correct ? | 2 Yes | 16 No |
why the most API molecule are in the hydrochloride form?
Why 1.2% KCl is used for limit of stray light? and why limit of stray light is required in UV calibration?
what does 'n' stands for in spectroscopy?give the formula.
in each media we require to use SLS? how to proceed?
Weight balance variation limit during analysis
7 Answers Amneal Pharmaceuticals,
why 0.005 M sulphuric acid is used to calbrate spectrophotometer and why caffiene is used for calibrating hplc
2 Answers GVK, Torrent Pharma,
How to start the dissolution development for unknown tab?
what is the most common FT-IR source used for the MID&Near IR regions?
1 Answers Dr Reddys, Piramal Enterprises,
Optical Rotation*0.9477*100*200] Assay Of dextrose= ________________________________ 99.8* 200 why Factor=0.9477
1 Answers Claris Lifesciences, Inven Pharmaceuticals, Levo,
Why we more noise in PDA detector compared to UV detector?
Nitrogen test is used for which chromatography?
Which parameter require to do for analytical method equivalency?