how to hplc calibrated
Answers were Sorted based on User's Feedback
HPLC is calibrated by the following parameters:
1) Pump calibration
2) Detector Linearity
3)Injector Precision and carryover test
4) Wavelength accuracy by using caffeine and pyrene solution
5) Determine the column oven temp. and sample compartment temp
6) Gradient test
7) Detector noise
| Is This Answer Correct ? | 128 Yes | 13 No |
Answer / ajinkya p.
In HPLC there must calibrations for Pump, Detector,injector
Pump :- Should calibrated by checking flow accuracy test, Pressure holding test.
Detector :- Should calibrated by checking Linearity test, Detector wavelength accuracy test
Auto sampler :- Should calibrated by Injector Accuracy test,
Carryover test, Injector precision test, carry over test
| Is This Answer Correct ? | 48 Yes | 2 No |
Answer / kanakadurga
calibration parameters are
1.Flow accuracy
2Temperature accuracy
3Wave length accuracy by using caffeine and erbium perchlorate
4Linearity of detector and injector
5.Carry over test
6.Drift and noice
| Is This Answer Correct ? | 37 Yes | 6 No |
Answer / b.ganesh
calibration parameters are
1.Flow accuracy
2Temperature accuracy
3Wave length accuracy by using caffeine and erbium perchlorate
4Linearity of detector and injector
5.Carry over test
6.Drift and noice
7.Gradient progromme accuracy
8.injector precision
| Is This Answer Correct ? | 3 Yes | 1 No |
Why we select scan range from higer wavelength to lower wavelength in uv visible spectroscopy and ftir spectroscopy ? .
what is Beer Lambert low ?
What is difference b/t uv detector and rid
is it necessary to do HPLC calibration at wavelength 315nm if we are doing analysis at this wavelangth
What is Formula of Potency calculation on Dried basis, As is basis& Anhydrous basis.
how to Calibrate FID IN GC
Why do we use 500ppb sucrose and 500ppb 1,4 benzoquinone in TOC calibration?
what is separation techniques.
(0.41)d what is d?
How much sample concentration should i inject against diluted standard in related substances for drug product in hplc analysis
What is Rf value& Rx value in TLC?
How did you conclude that when stressing a compond, all degradation products have eluted in chromatogram?