Why we are using K2Cr2O7 for control of absorbance in UV?
Answers were Sorted based on User's Feedback
Answer / smitadeb86
Control of absorbance is verified using known amount of Potassium dichromate solution and checking the absorbances at various wavelengths, to establish that the resulting E ( 1% 1 cm) levels at those wavelengths, are within the permitted tolerances.
| Is This Answer Correct ? | 8 Yes | 0 No |
Answer / d.lodhiya
Because its having 2 peak and 2 vallay . stable and
standard upto 3 hrs. its obays bees law.(ABS .25 to 1A)
| Is This Answer Correct ? | 9 Yes | 3 No |