Golgappa.net | Golgappa.org | BagIndia.net | BodyIndia.Com | CabIndia.net | CarsBikes.net | CarsBikes.org | CashIndia.net | ConsumerIndia.net | CookingIndia.net | DataIndia.net | DealIndia.net | EmailIndia.net | FirstTablet.com | FirstTourist.com | ForsaleIndia.net | IndiaBody.Com | IndiaCab.net | IndiaCash.net | IndiaModel.net | KidForum.net | OfficeIndia.net | PaysIndia.com | RestaurantIndia.net | RestaurantsIndia.net | SaleForum.net | SellForum.net | SoldIndia.com | StarIndia.net | TomatoCab.com | TomatoCabs.com | TownIndia.com
Interested to Buy Any Domain ? << Click Here >> for more details...


In the test for control of absorbance in UV calibration why
do we use only potassium dichromate and what is purpose of
taking a specified amount of 57.0-63.0mg?

Answers were Sorted based on User's Feedback



In the test for control of absorbance in UV calibration why do we use only potassium dichromate and..

Answer / manikatna

potassium dichromate is highly colourproducing agent
this amount is as per EP And edqm monograph.
if you change concentration change specific abs values

Is This Answer Correct ?    78 Yes 17 No

In the test for control of absorbance in UV calibration why do we use only potassium dichromate and..

Answer / bhagya lakshmi.k

It will show max absorbance in Ultraviolet region (200-
400nm) (i.e) at 235,257,313,350 nm

Is This Answer Correct ?    53 Yes 4 No

In the test for control of absorbance in UV calibration why do we use only potassium dichromate and..

Answer / sunil.m.kulkarni ( avon/arch p

Potassium Dichromate - UV Absorbance and Linearity
Product Description:
The use of potassium dichromate solvated in dilute
perchloric acid or suplhuric acid is an established and
well recognised method for the validation of the absorbance
scale and linearity of a spectrophotometer in the UV
region.
When prepared in 0.001M perchloric /Suplhuric acid,
potassium dichromate solution gives a spectral scan
containing characteristic peaks at 257 nm and 350 nm, and
troughs at 235 nm and 313 nm. Within the concen­tration
range 20-100 mg/l, if the absorbance scale of a narrow SBW
(< 2 nm) spectro­photometer is linear, the apparent
absorbances of a series of concentrations will be a linear
function of concentration.
Note: Primary standard ( Extra pure potassium dichromte
after drying to be used for calibration ) 57-63mg in -
1000ml concentration is perfect concentration for achiving
perfect charecteristic peak and throughs

Is This Answer Correct ?    36 Yes 5 No

In the test for control of absorbance in UV calibration why do we use only potassium dichromate and..

Answer / sagadevan

becoz of photochemical stability,so on prolonged exposure to uv light the absorbance own not change,

Is This Answer Correct ?    18 Yes 4 No

In the test for control of absorbance in UV calibration why do we use only potassium dichromate and..

Answer / devidas

it show the maxima at 235nm,257nm,313nm, and 350nm in UV
region. and it consider as a primary standard which is high
purity

Is This Answer Correct ?    18 Yes 7 No

In the test for control of absorbance in UV calibration why do we use only potassium dichromate and..

Answer / vivek

we are use potassium dichromate for the verification of spectrophotometer because it show maximum absorbtion in uv region...

Is This Answer Correct ?    12 Yes 5 No

In the test for control of absorbance in UV calibration why do we use only potassium dichromate and..

Answer / srinathsagar

UV range is 190-400nm so in this range potassium dichromate with sulfuric acid gives characteristic peaks(At 235,257,313 and 350nm)

Is This Answer Correct ?    4 Yes 1 No

In the test for control of absorbance in UV calibration why do we use only potassium dichromate and..

Answer / pvinod6447

because it covers uv to visible region i.e. 235nm,257nm,313nm,350nm

Is This Answer Correct ?    6 Yes 4 No

Post New Answer

More Analytical Chemistry Interview Questions

why using factor 1.00639 in burette calibration of kf

1 Answers   Cipla,


Defination of ZEEMAN EFFECT in NMR ?

2 Answers  


How can we determine Moisture content of HI?

1 Answers   Cipla, Reddy Labs,


is it necessary to do HPLC calibration at wavelength 315nm if we are doing analysis at this wavelangth

0 Answers   Aarti Drugs,


What is post run in hplc?

1 Answers   Cipla,


which effect is mainly responsible for chromatographic separation when silica is the statinary phase at TLC? thanks in advance

1 Answers  


what is the moisture content of Disodium hydrogen tartrate. how it can be calculted. as 16.5%

1 Answers  


In the test for control of absorbance in UV calibration why do we use only potassium dichromate and what is purpose of taking a specified amount of 57.0-63.0mg?

1 Answers   Reddy Labs,


why we can use only 4 and 7 and 9 buffers in pH calibration

1 Answers   Chola, Cipla,


What are the differences between GC and HPLC?

1 Answers  


If we have 5 strength which is not dose proportinate and excipients also diffrent in each strength then how we can proceed for Force degradation? and excipient are same but not dose propotinate the how FD?

0 Answers  


India is. Under which zone as per QF1.?

4 Answers   Acebright, Aurobindo, Gland Pharma,


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)