In the test for control of absorbance in UV calibration why
do we use only potassium dichromate and what is purpose of
taking a specified amount of 57.0-63.0mg?
Answers were Sorted based on User's Feedback
Answer / bhagya lakshmi.k
It will show max absorbance in Ultraviolet region (200-
400nm) (i.e) at 235,257,313,350 nm
Is This Answer Correct ? | 53 Yes | 4 No |
Answer / sunil.m.kulkarni ( avon/arch p
Potassium Dichromate - UV Absorbance and Linearity
Product Description:
The use of potassium dichromate solvated in dilute
perchloric acid or suplhuric acid is an established and
well recognised method for the validation of the absorbance
scale and linearity of a spectrophotometer in the UV
region.
When prepared in 0.001M perchloric /Suplhuric acid,
potassium dichromate solution gives a spectral scan
containing characteristic peaks at 257 nm and 350 nm, and
troughs at 235 nm and 313 nm. Within the concentration
range 20-100 mg/l, if the absorbance scale of a narrow SBW
(< 2 nm) spectrophotometer is linear, the apparent
absorbances of a series of concentrations will be a linear
function of concentration.
Note: Primary standard ( Extra pure potassium dichromte
after drying to be used for calibration ) 57-63mg in -
1000ml concentration is perfect concentration for achiving
perfect charecteristic peak and throughs
Is This Answer Correct ? | 36 Yes | 5 No |
Answer / sagadevan
becoz of photochemical stability,so on prolonged exposure to uv light the absorbance own not change,
Is This Answer Correct ? | 18 Yes | 4 No |
Answer / devidas
it show the maxima at 235nm,257nm,313nm, and 350nm in UV
region. and it consider as a primary standard which is high
purity
Is This Answer Correct ? | 18 Yes | 7 No |
Answer / vivek
we are use potassium dichromate for the verification of spectrophotometer because it show maximum absorbtion in uv region...
Is This Answer Correct ? | 12 Yes | 5 No |
Answer / srinathsagar
UV range is 190-400nm so in this range potassium dichromate with sulfuric acid gives characteristic peaks(At 235,257,313 and 350nm)
Is This Answer Correct ? | 4 Yes | 1 No |
because it covers uv to visible region i.e. 235nm,257nm,313nm,350nm
Is This Answer Correct ? | 6 Yes | 4 No |
What is the defination of Bulk density?
For standardization of volumetric solutions, acceptance limit for concentration is considered as 10% of nominal concentration and %RSD as NMT 0.2% (for appropriate no. of replicates). But , if %RSD has not met, what should be further action?
What is Specificity ?
In Assy by potentio metry titration method how select MET U ,DET U, SET U,
Which type of material analysis in GC using Theramal conditivity detector ?
defination of wavelength & wave number
why we are analyzing purity on HPLC why not in gc
3 Answers Aurobindo, Reddy Labs,
What is Signal to noise ratio?
Why using 0.5% acetone in methanol mobile phase for HPLC Gradient performance check.
When we calculate Impurity against impurity std.solution and when we calculate impurity against diluted standard with applying RRF?
Explain in detail what are the procedure and calculations involved for accuracy (%Recovery)in method validation?
why kf factor coming 0.5