In the test for control of absorbance in UV calibration why
do we use only potassium dichromate and what is purpose of
taking a specified amount of 57.0-63.0mg?
Answers were Sorted based on User's Feedback
Answer / bhagya lakshmi.k
It will show max absorbance in Ultraviolet region (200-
400nm) (i.e) at 235,257,313,350 nm
| Is This Answer Correct ? | 53 Yes | 4 No |
Answer / sunil.m.kulkarni ( avon/arch p
Potassium Dichromate - UV Absorbance and Linearity
Product Description:
The use of potassium dichromate solvated in dilute
perchloric acid or suplhuric acid is an established and
well recognised method for the validation of the absorbance
scale and linearity of a spectrophotometer in the UV
region.
When prepared in 0.001M perchloric /Suplhuric acid,
potassium dichromate solution gives a spectral scan
containing characteristic peaks at 257 nm and 350 nm, and
troughs at 235 nm and 313 nm. Within the concentration
range 20-100 mg/l, if the absorbance scale of a narrow SBW
(< 2 nm) spectrophotometer is linear, the apparent
absorbances of a series of concentrations will be a linear
function of concentration.
Note: Primary standard ( Extra pure potassium dichromte
after drying to be used for calibration ) 57-63mg in -
1000ml concentration is perfect concentration for achiving
perfect charecteristic peak and throughs
| Is This Answer Correct ? | 36 Yes | 5 No |
Answer / sagadevan
becoz of photochemical stability,so on prolonged exposure to uv light the absorbance own not change,
| Is This Answer Correct ? | 18 Yes | 4 No |
Answer / devidas
it show the maxima at 235nm,257nm,313nm, and 350nm in UV
region. and it consider as a primary standard which is high
purity
| Is This Answer Correct ? | 18 Yes | 7 No |
Answer / vivek
we are use potassium dichromate for the verification of spectrophotometer because it show maximum absorbtion in uv region...
| Is This Answer Correct ? | 12 Yes | 5 No |
Answer / srinathsagar
UV range is 190-400nm so in this range potassium dichromate with sulfuric acid gives characteristic peaks(At 235,257,313 and 350nm)
| Is This Answer Correct ? | 4 Yes | 1 No |
because it covers uv to visible region i.e. 235nm,257nm,313nm,350nm
| Is This Answer Correct ? | 6 Yes | 4 No |
What is the difference between ROI & sulphates ash?
how to regulate noise in hplc ?
why we are adjusting the ph in buffer solution preparation
In acid base titration, after complet consumption of HCl by NaOH, addition of next drop of NaOH will give pink colour with phenolpthalein indicator. Why this pink colour appear? Why not blue, black, green or anything elese?
in monograph 7 imp are there but in vendor showing 5 imp do not req to include 2 imp for that what will be justification?
control of resolution in calibration of IR
What is the SOP for SOP
Calculate the water content in Di Sodium Tartrate (DST) based on molecular weight?
Hplc detect how much ppm solutions i want range
what is the difference between polarimeter lamp and ir lamp
in hplc calibration injector linearity pass,it means inj and det is good. why to perform detector linearity is it required, if it is required,what is difference between injector and detector linearity in hplc calibration
in hplc calibration we are using 0.05% acetone why we are using that one and what happen when its conc increases or decreases
4 Answers Aurobindo, Cadila, Martin,