In the test for control of absorbance in UV calibration why
do we use only potassium dichromate and what is purpose of
taking a specified amount of 57.0-63.0mg?

Answers were Sorted based on User's Feedback



In the test for control of absorbance in UV calibration why do we use only potassium dichromate and..

Answer / manikatna

potassium dichromate is highly colourproducing agent
this amount is as per EP And edqm monograph.
if you change concentration change specific abs values

Is This Answer Correct ?    78 Yes 17 No

In the test for control of absorbance in UV calibration why do we use only potassium dichromate and..

Answer / bhagya lakshmi.k

It will show max absorbance in Ultraviolet region (200-
400nm) (i.e) at 235,257,313,350 nm

Is This Answer Correct ?    53 Yes 4 No

In the test for control of absorbance in UV calibration why do we use only potassium dichromate and..

Answer / sunil.m.kulkarni ( avon/arch p

Potassium Dichromate - UV Absorbance and Linearity
Product Description:
The use of potassium dichromate solvated in dilute
perchloric acid or suplhuric acid is an established and
well recognised method for the validation of the absorbance
scale and linearity of a spectrophotometer in the UV
region.
When prepared in 0.001M perchloric /Suplhuric acid,
potassium dichromate solution gives a spectral scan
containing characteristic peaks at 257 nm and 350 nm, and
troughs at 235 nm and 313 nm. Within the concen­tration
range 20-100 mg/l, if the absorbance scale of a narrow SBW
(< 2 nm) spectro­photometer is linear, the apparent
absorbances of a series of concentrations will be a linear
function of concentration.
Note: Primary standard ( Extra pure potassium dichromte
after drying to be used for calibration ) 57-63mg in -
1000ml concentration is perfect concentration for achiving
perfect charecteristic peak and throughs

Is This Answer Correct ?    36 Yes 5 No

In the test for control of absorbance in UV calibration why do we use only potassium dichromate and..

Answer / sagadevan

becoz of photochemical stability,so on prolonged exposure to uv light the absorbance own not change,

Is This Answer Correct ?    18 Yes 4 No

In the test for control of absorbance in UV calibration why do we use only potassium dichromate and..

Answer / devidas

it show the maxima at 235nm,257nm,313nm, and 350nm in UV
region. and it consider as a primary standard which is high
purity

Is This Answer Correct ?    18 Yes 7 No

In the test for control of absorbance in UV calibration why do we use only potassium dichromate and..

Answer / vivek

we are use potassium dichromate for the verification of spectrophotometer because it show maximum absorbtion in uv region...

Is This Answer Correct ?    12 Yes 5 No

In the test for control of absorbance in UV calibration why do we use only potassium dichromate and..

Answer / srinathsagar

UV range is 190-400nm so in this range potassium dichromate with sulfuric acid gives characteristic peaks(At 235,257,313 and 350nm)

Is This Answer Correct ?    4 Yes 1 No

In the test for control of absorbance in UV calibration why do we use only potassium dichromate and..

Answer / pvinod6447

because it covers uv to visible region i.e. 235nm,257nm,313nm,350nm

Is This Answer Correct ?    6 Yes 4 No

Post New Answer

More Analytical Chemistry Interview Questions

What is the defination of Bulk density?

4 Answers   Dr Reddys,


For standardization of volumetric solutions, acceptance limit for concentration is considered as 10% of nominal concentration and %RSD as NMT 0.2% (for appropriate no. of replicates). But , if %RSD has not met, what should be further action?

0 Answers  


What is Specificity ?

8 Answers   Dr Reddys,


In Assy by potentio metry titration method how select MET U ,DET U, SET U,

0 Answers  


Which type of material analysis in GC using Theramal conditivity detector ?

7 Answers  






defination of wavelength & wave number

4 Answers  


why we are analyzing purity on HPLC why not in gc

3 Answers   Aurobindo, Reddy Labs,


What is Signal to noise ratio?

3 Answers  


Why using 0.5% acetone in methanol mobile phase for HPLC Gradient performance check.

2 Answers   DRL,


When we calculate Impurity against impurity std.solution and when we calculate impurity against diluted standard with applying RRF?

2 Answers   Torrent Pharma,


Explain in detail what are the procedure and calculations involved for accuracy (%Recovery)in method validation?

1 Answers   Vimta Labs,


why kf factor coming 0.5

5 Answers   Alkem Labs,


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)