Answer / m
Specificity is proves that your method of analysis is
capable or not to resolve the all know impurities, unknown
impurities, degradation product and matrix compound by
proving the peak purity and also prove that all known
impurity are well separate from each other.
| Is This Answer Correct ? | 50 Yes | 1 No |
The specificity of an analytical method is basically to
1) To prove the stability indicating ability of the method
2)To identify the possible degradation of the molecule
3)To determine the type of degradation the molecule is
susceptible to
4)To prove the mass balance
5)To know the various degradation path ways of molecule
| Is This Answer Correct ? | 11 Yes | 2 No |
Answer / nitin
specipicity is do to known whether any excipient interfere
with our analyte drug peak resolusion .
| Is This Answer Correct ? | 10 Yes | 1 No |
Answer / sachin.bedwal
IN SPECIFICITY WE HAVE TO IDENTYFY THE CONCENTRAION OF
SPECIFIC COMPONENT IN ENTIRE MIXTURE
| Is This Answer Correct ? | 6 Yes | 2 No |
Answer / santosh
Specificity means spefic to main analyte whether that main analyte is separated properly or not with out any coelution with any other impurities no need to check all impurities
| Is This Answer Correct ? | 0 Yes | 0 No |
What is chiral and why this performing in HPLC
What is the exact reason to use THF and IPA in buffers of Reverse phase HPLC as a solvent?
cefoperozone and sulbactam inj. hplc test method
0 Answers Graviti Pharmaceuticals,
in karl fischer reagent which solvent is used for dissolving iodine sulphur dioxide and imidazole
full name of OOS & OOT ? Dercribe it with details ? why this is needed in pharma industry ?
how can we decide the limit of cleaning validation? What is the criteria for it?
For standardization of volumetric solutions, acceptance limit for concentration is considered as 10% of nominal concentration and %RSD as NMT 0.2% (for appropriate no. of replicates). But , if %RSD has not met, what should be further action?
iam using ph instrument lab india make.3ponit and 5point.in calibration time 3point slope will come 98.89and 96.87.that limit is 85 to 105.what basis slope will come.any calucalation between them.5point wiicome slope 99.89?why 3p0int came 2 values
WHAT IS USE OF ION PAIR REAGENTS?
what is the origin to prepare standard operating procedure
why we are taking 85 ml of concentrated HCl to prepare 1N Solution
why sucrose used in polarimeter calibration?
2 Answers Reddy Labs, Sugar Factory,