What is Delay & Dead Volume
Answers were Sorted based on User's Feedback
Answer / rahul shimpi
DELAY volume; the volume from the GPV TO DETECTOR AND DEAD
VOLUME IS THE VOLUME FORM SAMPLE INJECTOR TO COLUMN OUTLET
Is This Answer Correct ? | 24 Yes | 5 No |
Answer / rahul shimpi
gpv to detector is the dwell volume. delay is the volume
taken extra by the system to start the gradient. suppose
inm gradient starts at 4.0 min it will not start at 4.0min
instead of that it will start at 4.5 min so 0.5 min delay
Is This Answer Correct ? | 18 Yes | 4 No |
Why we use holmium oxide,perchloric acid for uv/vis calibration
4 Answers Cipla, Orchid, Ratiopharm,
What do you mean by potency in Analytical industry
Why dissolution test apparatus calibration with salycilic acid tablets was stopped by FDA&USP from the year of 2009?
What is cleaning validation of rinse and swabe and its limit with caculation formula.
Tell me about analytical method validation in QC
Why we are performing wavelength accuracy at lower and higher wavelengths in HPLC calibration
What is third generation HPLC Columns?
What is colimn selection Criteria in HPLC analysis?
How do we get end points and how many end points are possible for citric acid and di-acid not theorotically answer should be given practically.
How require to perform linearity as per ANVISA? What are the acceptance criteria?
What is the difference between Optical rotation and Spcefic optical rotation
what is the acceptance criteria for enteric coated tablets in 0.1n hcl validation in each parameter?