control of resolution in calibration of IR
Answers were Sorted based on User's Feedback
Answer / nirav intas pharma
control of resolution in IR:
Take spectrum of polystyrine film,
the abs minima at 2870 cm- & abs maxima at 2849.5cm- is
greater than 0.33
the abs minima at 1589 cm- & abs maxima at 1583 cm- is
greater than 0.08
| Is This Answer Correct ? | 12 Yes | 0 No |
Answer / kuchela rao athota
MAXIMUM ABSORPTION AT ABPUT 1583CM-1 SHOULD BE GREATER THAN
12%
MAXIMUM ABSORPTION AT ABPUT 2850CM-1 SHOULD BE GREATER THAN
18%
| Is This Answer Correct ? | 5 Yes | 0 No |
Answer / arpit vashi
the difference between two percentage transmittance
(2870 cm-1 and 2849.5 cm-1) must be grater than 18%
the difference between two percentage transmittance
(1589 cm-1 and 1583 cm-1) must be grater than 10%
| Is This Answer Correct ? | 1 Yes | 0 No |
WHAT IS THE CRITERIA FOR SELECTION OF TIME OF DISSOLUTION AND THE MEDIUM OF DISSOLUTION?
What is the Formula for coreletion coefficient in plhplc calibration
how to correct peak when peak comes broader in shape? How to do calibration of HPLC?
what is difirence b/w ROI&ROE
In pharmacopeal solibility parameter of a material, is et essential to test all paramete? If 70-75% parameter passes can we assess that the material is satisfactory or not?
Why we use caffeine for HPLC calibration in Detector wave ength accuracy test
60 Answers Arvind, Aurobindo, Cipla, Reddy Labs, Shasun Chemicals,
In which situation we require to analytical method validation of excipient?
what is MCG dilution in HPLC & GC analysis
1 Answers Amoli Organics, Cipla,
Why we are use 0.005M Sulphuric acid, in control of absorbance performance test UV Spectrophotometer calibration
define beer-lambertslaw.
How to fix the sample conc. in Related substances test by HPLC?
principle of AAS & its uses ?