control of resolution in calibration of IR
Answers were Sorted based on User's Feedback
Answer / nirav intas pharma
control of resolution in IR:
Take spectrum of polystyrine film,
the abs minima at 2870 cm- & abs maxima at 2849.5cm- is
greater than 0.33
the abs minima at 1589 cm- & abs maxima at 1583 cm- is
greater than 0.08
| Is This Answer Correct ? | 12 Yes | 0 No |
Answer / kuchela rao athota
MAXIMUM ABSORPTION AT ABPUT 1583CM-1 SHOULD BE GREATER THAN
12%
MAXIMUM ABSORPTION AT ABPUT 2850CM-1 SHOULD BE GREATER THAN
18%
| Is This Answer Correct ? | 5 Yes | 0 No |
Answer / arpit vashi
the difference between two percentage transmittance
(2870 cm-1 and 2849.5 cm-1) must be grater than 18%
the difference between two percentage transmittance
(1589 cm-1 and 1583 cm-1) must be grater than 10%
| Is This Answer Correct ? | 1 Yes | 0 No |
what is the limit of Correlation for IR test compare to the standard and test
why inject diluted standard solution in related substance analysis by HPLC?
How to choose concentration of LOQ solutions for RS analysis? Example.
1 Answers Abbott, Indoco Remedies,
USP methodology, EP methodology, IP methodology, among three if possible to use one methodology for qualify working standard to use USP, EP, IP ? Please explain...
0 Answers Srini Pharmaceuticals,
what is the difference between system precision and method precision?
3 Answers Teva Pharmaceuticals,
in hplc chromatogram started from left to right and in uv spectrum started from left to right
how pyridine can be analysed other than GC?
When we calculate Impurity against impurity std.solution and when we calculate impurity against diluted standard with applying RRF?
what is the differance between Analytical and semi micro balance?
why we multiply 1000 in the determination of kf factor ?
3 Answers Stellence Pharm Science,
why we inject sample TWICE in ASSAY by HPLC analysis but in RS once?
4 Answers Aurobindo, DRL, Neuland,
wt is rsd in and explain