control of resolution in calibration of IR
Answers were Sorted based on User's Feedback
Answer / nirav intas pharma
control of resolution in IR:
Take spectrum of polystyrine film,
the abs minima at 2870 cm- & abs maxima at 2849.5cm- is
greater than 0.33
the abs minima at 1589 cm- & abs maxima at 1583 cm- is
greater than 0.08
| Is This Answer Correct ? | 12 Yes | 0 No |
Answer / kuchela rao athota
MAXIMUM ABSORPTION AT ABPUT 1583CM-1 SHOULD BE GREATER THAN
12%
MAXIMUM ABSORPTION AT ABPUT 2850CM-1 SHOULD BE GREATER THAN
18%
| Is This Answer Correct ? | 5 Yes | 0 No |
Answer / arpit vashi
the difference between two percentage transmittance
(2870 cm-1 and 2849.5 cm-1) must be grater than 18%
the difference between two percentage transmittance
(1589 cm-1 and 1583 cm-1) must be grater than 10%
| Is This Answer Correct ? | 1 Yes | 0 No |
How is the Flame test performed?
principle of FTIR
what is main use of primary standards in caliration
2 Answers Lee Pharma, Lupin, Nectar Lifesciences,
if your product is soluble in 0.1n hcl and water then which you choose as media from these 2 media?
How many clauses in ISO!
what is the difference between EP and BP
2 Answers Novartis, Vimta Labs,
which brands GC coloumn available in market
In UV calibration we use Holmium perchlorate solution made in 1.4M perchloric acid The question is that 1.4 M perchloric acid we make is it in water or glacial acetic acid?
how to selecet an exact coloumn for an new molecule development by hplc how to select exact salt as buffer for new molecule development by hplc what is the the process to select the mode of saparation of compoundes by hplc what is the use of ph of buffer what is use of buffer,ph,organic phase,ans methods how the molecules get saparated in coloumn,
How to do spike of aflatoxin in peanut to get maximum recovery
Why do we use 500ppb sucrose and 500ppb 1,4 benzoquinone in TOC calibration?
If suppose 10 methods of dissolution given in pharmacopoeial for single content product then which method out of 10? or all 10 require to follow?