In thermocouple could we use the lug for termination or
not,Thermocouple is any time,I am currently concerned with
K type.please let me know if anybody know
Answers were Sorted based on User's Feedback
Answer / jigaranadkat
LAW of intermediate metal comes into picture, according to
which resulting temperature will be the algebric sum of all
EMF produced inbetween.
But with LUGs perticularly, the temperature difference will
be very less between so less almost negligible amount of
EMF will be produced. Hence NO HARm.
In order to achieve high accuracy it has been observed that
perticularly in ares where temperatures are TOO HIGH,
LUGing is to be avoided.
| Is This Answer Correct ? | 11 Yes | 0 No |
Answer / george mathews
The basic principle behind any Thermocouple is that there
are two disimilar material connected at one end that is
called the Hot Junction , and we take our out put from the
other open junction that is the cold junction. But when we
connect a lug there is anather metal thats mean 3 metal and
we ll definitely get a wrong out put and in a high
temperature area or zone this may make a big difference.
| Is This Answer Correct ? | 14 Yes | 5 No |
Answer / bm
Dont see any harm in connecting lug for termination
provided the continuity remains the same i.e there will be
very very less resistance drop if proper connection is not
there with thermocouple wire, lug and the terminals.
| Is This Answer Correct ? | 6 Yes | 5 No |
Answer / rathina
Normal practice, Leg connection not recommended
| Is This Answer Correct ? | 1 Yes | 2 No |
if I did short to both AO connections, channel will burn or what? m sure in AI it will damage to channel. But what will happens in AO?
what is the difference b/w normal thermocouple and skin type thermocouple?
what is the meaning of reverse action and forward action of control valve...
9 Answers Honeywell, Yokogawa,
What is upsampling and downsampling?
Somebody tell me in what way i need to prepare for csio pgrpe(adv instrum eng)interview?what all the quest can i expect?
generally transmitter power supply having some range(12v to 35v) but the out put(4-20 mA) must same for given input.If we vary the power supply voltage then output mA also to be change according to V=IR. Is there any voltage regulator in the transmitter ?
what is the deference bitween DCS and PLC SCADA HMI System for Power Process Industry.
An engineering term STALL used in compressor operation what it means?
hello friends, can you please send me the model paper or question paper to be prepared for hpcl examinations????????? You can mail me at rahim_here@yahoo.com waiting for your reply........... help me .. plz.....
Can Any body tell me how to check Modbus Settings in ER300N L&T trivector meter?
Thermocouple is a transducer but rtd is not transducer why? Both taking heat as a input and giving mv
is here anybody who knows how to calibrate the pressure transmitter? ONLINE
Civil Engineering (5086)
Mechanical Engineering (4453)
Electrical Engineering (16638)
Electronics Communications (3918)
Chemical Engineering (1095)
Aeronautical Engineering (239)
Bio Engineering (96)
Metallurgy (361)
Industrial Engineering (259)
Instrumentation (3014)
Automobile Engineering (332)
Mechatronics Engineering (97)
Marine Engineering (124)
Power Plant Engineering (172)
Textile Engineering (575)
Production Engineering (25)
Satellite Systems Engineering (106)
Engineering AllOther (1379)