Golgappa.net | Golgappa.org | BagIndia.net | BodyIndia.Com | CabIndia.net | CarsBikes.net | CarsBikes.org | CashIndia.net | ConsumerIndia.net | CookingIndia.net | DataIndia.net | DealIndia.net | EmailIndia.net | FirstTablet.com | FirstTourist.com | ForsaleIndia.net | IndiaBody.Com | IndiaCab.net | IndiaCash.net | IndiaModel.net | KidForum.net | OfficeIndia.net | PaysIndia.com | RestaurantIndia.net | RestaurantsIndia.net | SaleForum.net | SellForum.net | SoldIndia.com | StarIndia.net | TomatoCab.com | TomatoCabs.com | TownIndia.com
Interested to Buy Any Domain ? << Click Here >> for more details...


How can you calculate manually "Slope" in pH calibration?

Answers were Sorted based on User's Feedback



How can you calculate manually "Slope" in pH calibration?..

Answer / chmreddy

A good electrode has a mV of 0 ± 30 mV in pH 7.
The slope should be 90 to 105%.
Example:
Readings
pH 4 : 4.29
pH 7 : 7.43
Convert pH 7 reading to mV
7 – 7.43 = -0.43 pH * 59.16 mV/pH = -25.44 (this value is okay because it is within 0 ± 30mV)
Convert pH 4 reading to mV
7 – 4.29 = 2.71 pH * 59.16 mV/pH = 160.32
Calculate slope
160.32 – (-25.44) = 185.76 mV/ 177.48 mV * 100 = 105 %
(this value is borderline as 105% is the upper limit)

Is This Answer Correct ?    13 Yes 11 No

How can you calculate manually "Slope" in pH calibration?..

Answer / narendra jha

u calculate slope of pH against pH 7

Is This Answer Correct ?    4 Yes 5 No

Post New Answer

More Analytical Chemistry Interview Questions

if tech transfer fails whay require to do?

0 Answers  


For standardization of volumetric solutions, acceptance limit for concentration is considered as 10% of nominal concentration and %RSD as NMT 0.2% (for appropriate no. of replicates). But , if %RSD has not met, what should be further action?

0 Answers  


principle of FTIR

2 Answers   Biocon, Macleods,


what is the significant of pH in HPlc?

5 Answers   Orchid,


How to know the estimated LOQ concentration in ppm

0 Answers  


In Stability testing if significant change occurs then what will be the action plan?

2 Answers   Cipla,


How can you prepare 0.05M Ammonium acetate volumetric solution?

2 Answers  


IS THERE ANY EQUIPMENT TO CHECK AND CALCULATE THE POLARITY OF A LIQUID?

0 Answers   Reddy Labs,


Why do we use KMnO4 in the test of control of obsorbance ? and why do we take specific quantity i.e 57-63mg?

0 Answers   Reddy Labs, Vimta Labs,


In HPLC (Waters) software Empower1,2 &3 what is the updation for each version

0 Answers   APL,


what is the difference between the ordinary loss on drying technique and vacuum oven technique?

1 Answers   Divis Laboratories,


how to develop the icp ms method? Application of icp ms?

0 Answers  


Categories
  • Organic Chemistry Interview Questions Organic Chemistry (342)
  • Inorganic Chemistry Interview Questions Inorganic Chemistry (123)
  • Analytical Chemistry Interview Questions Analytical Chemistry (1387)
  • Physical Chemistry Interview Questions Physical Chemistry (62)
  • General Chemistry Interview Questions General Chemistry (478)
  • Chemistry AllOther Interview Questions Chemistry AllOther (190)