in hplc calibration injector linearity pass,it means inj and det is good. why to perform detector linearity is it required, if it is required,what is difference between injector and detector linearity in hplc calibration
Answers were Sorted based on User's Feedback
Answer / sudhakar tambe (gangapur)
injector linearity is perform for accuracy of injector or syringe volume accuracy. and detector linearity perform for got the same res ponce of area at specified wavelength.
| Is This Answer Correct ? | 14 Yes | 0 No |
Answer / ch.murali krishna
In detector lineariry inj vol same but different conc to check the detector performance of the system
In inj linearity conc same inj vol different to check the linearity of injection 💉
| Is This Answer Correct ? | 5 Yes | 0 No |
Answer / bhanumurthy.s
Both are required.In Injection volume linearity injection volume changed where as in Detector linearity constant injection volume with different Concentrations.
| Is This Answer Correct ? | 4 Yes | 1 No |
Answer / dave
but Bhanumurthy.S sir
during the injector linearity performance test, as the volume change the concentration also change, it is done by system
where as in the detector linearity we manually change the concentration so ultimately we also perform detector linearity during the injector linearity test.
| Is This Answer Correct ? | 0 Yes | 3 No |
Suppose we prepared pH buffer solution in lab.and calibrate against NIST solution, and same shall going to used upto 30 days. is it possible to stable pH and apperance ? suggest ?
Why used 0.02% v/v solution of toluene in hexane for calibration of UV.
are there in uv lamb that is chractrestic for wavelenghth 205 nm ?????
1 Answers Beni Suef University, Graviti Pharmaceuticals, Sri Krishna Pharmaceuticals,
What is aggregate and fragments in SEC?
What is primary & Secondary Standard ? Give Example
As per USP u/vacuum means (how much vacuum should be applied)?
I want analyze my compound at 325 nm so which standard shall i use to calibrate my HPLC system? How i can assure my hplc detector is suitable or not?
WHAT IS DIFFERENCE BETWEEN RRT AND RRF IN HPLC
which are the sizes of capsules?
In UV calibration we use Holmium perchlorate solution made in 1.4M perchloric acid The question is that 1.4 M perchloric acid we make is it in water or glacial acetic acid?
In GC Why does retention time decrease when temperature increases? Also how does a change in temperature effect selectivity?
what is the Bracketing & Matrixing.