in hplc calibration injector linearity pass,it means inj and det is good. why to perform detector linearity is it required, if it is required,what is difference between injector and detector linearity in hplc calibration
Answers were Sorted based on User's Feedback
Answer / sudhakar tambe (gangapur)
injector linearity is perform for accuracy of injector or syringe volume accuracy. and detector linearity perform for got the same res ponce of area at specified wavelength.
| Is This Answer Correct ? | 14 Yes | 0 No |
Answer / ch.murali krishna
In detector lineariry inj vol same but different conc to check the detector performance of the system
In inj linearity conc same inj vol different to check the linearity of injection 💉
| Is This Answer Correct ? | 5 Yes | 0 No |
Answer / bhanumurthy.s
Both are required.In Injection volume linearity injection volume changed where as in Detector linearity constant injection volume with different Concentrations.
| Is This Answer Correct ? | 4 Yes | 1 No |
Answer / dave
but Bhanumurthy.S sir
during the injector linearity performance test, as the volume change the concentration also change, it is done by system
where as in the detector linearity we manually change the concentration so ultimately we also perform detector linearity during the injector linearity test.
| Is This Answer Correct ? | 0 Yes | 3 No |
what is the definition of limit of stray light in UV calibration
in karl fischer reagent which solvent is used for dissolving iodine sulphur dioxide and imidazole
why we measure PH of a solution in PH only why not POH ?
1 Answers Stellence Pharm Science,
how to prepare solution in ppb
class of solvent regarding Residual solvent e.g. Acetone his class
loss on drying we applied vaccum for some compounds and some compounds are at 105c why
why base line not starts from zero in gc?
1 Answers Aurobindo, FDC, Reddy Labs,
What is the pH of water?
Ratio of Water and Acetonitrile(diluent) is not recommended to use in cleaning method validation,why?
0 Answers Aurobindo, Reddy Labs,
what is the difference between loss on drying in ordinary oven technique and vacuum oven technique
4 Answers Divis Laboratories, Ranbaxy,
what is mean by Q in S1 and S2
Through assay by hplc(anhydrous basis) product got 100.8% .How much the potency of that product.Moisture content is 1.7%,Sulfated ash is 0.02%,Total impurities are 0.3%