Why we more noise in PDA detector compared to UV detector?
Answer / anil
There is no Integration function in the conventional UV-Vis spectrophotometer which
accumulates the signal. For example, the total required time will be 1000 sec. for 1000 data
points and it takes 1 sec. to measure one datum. In this case, all 1000 data have the same signal
to noise ratio (S/N).
But in a PDA instrument which has a 1000 photodiode array, 1000 data points can be
measured in 1 sec. and it would take 1/1000 sec. to achieve the same result obtainable in 1 sec.
in a conventional instrument. Therefore, when the same sample is measured for 1000 sec in a
PDA instrument, the signal is accumulated and is 1000 times greater than when measuring for 1
sec. The noise will be 10001/2
. This means that the S/N ratio is improved by 10001/2
.
This resulting benefit of fast data acquisition is termed Felgett’s S/N Advantage or
Multichannel Advantage
| Is This Answer Correct ? | 10 Yes | 1 No |
wt is difference between assay and purity in hplc?
which situation gc hs and gc als require to use?
effect of Ph during analytical run ?
1 Answers Lambda Therapeutic Research,
I have compare C2H2-air and C2H2-N2O flame AAS on determination calcium. Both use same range of std to plot calibration curve. (2-6ppm) When i measure the sample with phosphate, KCl and LaCl, C2H2-N2O flame give false positive result, around 0.5ppm. When i measure the sample with phosphste, KCl and EDTA. C2H2-N2O flame also give 0.5ppm false positive. But both above mentioned sample would not give false positive when measured by C2H2-air flame. What is the reason?
What is difference b/w IR&NIR
What impact analysis required for out of calibration and how to proceed
if you given one product then which tests you will perform?
WHAT IS DIFFERENCE B/W MALTINGRANG/DSC
Why red HgI2 added during starch preparation?
what is peak purity and its formula
define polar and non polar solvent
WHY pKa value is more important than pH value of the mobile phase in HPLC?