Why we more noise in PDA detector compared to UV detector?
Answer / anil
There is no Integration function in the conventional UV-Vis spectrophotometer which
accumulates the signal. For example, the total required time will be 1000 sec. for 1000 data
points and it takes 1 sec. to measure one datum. In this case, all 1000 data have the same signal
to noise ratio (S/N).
But in a PDA instrument which has a 1000 photodiode array, 1000 data points can be
measured in 1 sec. and it would take 1/1000 sec. to achieve the same result obtainable in 1 sec.
in a conventional instrument. Therefore, when the same sample is measured for 1000 sec in a
PDA instrument, the signal is accumulated and is 1000 times greater than when measuring for 1
sec. The noise will be 10001/2
. This means that the S/N ratio is improved by 10001/2
.
This resulting benefit of fast data acquisition is termed Felgett’s S/N Advantage or
Multichannel Advantage
| Is This Answer Correct ? | 10 Yes | 1 No |
Wt s the use of linearity in HPLC method validation
Can we interpret accuracy from linearity in method validation?
Why linearity of detector performed in HPLC calibration
calibrtion procedue of LC-MS
How can I develope method of dissolution by HPLC OR UV
How much area of methanol or other solvent are acceptable in blank (Diluent) of Gas chromatography.
we are performed SOR for a particular product the limit is -6 to -10.We face particular bathes it is not meeting the specification.we performed diffrent instrument diffrent analyst we are getting diffrent diffrent results.solvent is dimethyl sulfoxide what could be the reasion
2. Two grams of Benzoic acid are dissolved in 200 ml of water and extracted with 200 ml of diethyl ether. The distribution coefficient of benzoic acid is 100, and its dissociation constant is 6.5 10-5. Calculate the distribution ratio (D) of benzoic acid at pH 2, 5, and 6. 3. Calculate D at pH 2 to 10 (1 unit apart) in the above problem, and plot D versus pH.
1.What is the difference between method validation and method verification 2.Which guidelines proposed to method transfer
Why pH range is maximum 14.0
How to prepare 50 ppm ammonia solution form 25% ammonia solution?
how to develop the shorter run time for pharmacopeial methods having 100 min run time?