What is Specificity ?
Answers were Sorted based on User's Feedback
Answer / subash
Specificity is the ability to assess unequivocally the
analyte in the presence of components which might be
expected to be present. These might include impurities,
degradants, complex matrices, etc.
| Is This Answer Correct ? | 11 Yes | 0 No |
Answer / vidyaevuri
Specificity determines the consider peak is single or it co-
eluting with another peak with same retention time. And
peak purity should be pass for comsider peak.
There is no interference of balnk and placebo peaks with
Active peak and impurity peaks. also
| Is This Answer Correct ? | 9 Yes | 0 No |
Answer / phanindra
this parameter is used to check the purity's of sample and impurities which we are used in validation,Retention time identification,check the separation of impurities and also used to calculation of RRT AND RRf's
| Is This Answer Correct ? | 6 Yes | 1 No |
Answer / vaishali musale
The specificity of the method was determined by analyzing
standard drug &test samples.the purity of the drug
determined by comparing Three regions i.e. peak start peak
apex & peak end.
| Is This Answer Correct ? | 4 Yes | 3 No |
Answer / sham
Specificity is one of the experiment name during method
validation. As the name suggests, this exp is done to find
how specific the method is for the analytes being analysed.
PDA detector is used to ensure the absence of co-elusion
and ultimately the specificity.
| Is This Answer Correct ? | 2 Yes | 2 No |
Answer / eswar
If the drug substance not dissolved in dissolution medium,then what we do?
| Is This Answer Correct ? | 0 Yes | 0 No |
Answer / srinivasreddy
Specificity determines the consider peak is single or it co-eluting with another peak with same retention time. And
peak purity should be pass for consider peak.
There is no interference of balnk and placebo peaks with
Active peak and impurity peaks. also
| Is This Answer Correct ? | 0 Yes | 0 No |
what is difference b/w spectroscopy, Spectrometry and Spectrophotometry ?
HOW TO CALIBRATE HPLC OVEN
how do u optimise hplc method
What is Formula of Potency calculation on Dried basis, As is basis& Anhydrous basis.
What is heavy water
%lod always grater than %water content for same product??
diference b/w packed &capilary columns
iam using ph instrument lab india make.3ponit and 5point.in calibration time 3point slope will come 98.89and 96.87.that limit is 85 to 105.what basis slope will come.any calucalation between them.5point wiicome slope 99.89?why 3p0int came 2 values
what is the moisture content of Disodium hydrogen tartrate. how it can be calculted. as 16.5%
Tell me about analytical method validation in QC
why do we use around only 60 mg of K2cr2o7 in the UV calibration -control of absorbance test?
what is the limit for KF Factor ,we take factor using water we will found factor between 6 to 7 ,while on KF Reagent bottle mentioned 5 mg/ml water content.if we not use dried methanol it is effect on factor.