different between calibration range and instrument range ?
Answers were Sorted based on User's Feedback
Calibration range at which the
instrument calibrated
instrument range:maximum range of
instrument at which we can calibrate
instrument for example if we have a
pressure tx instrument range 0~10bar
we can calibrate it at any position
between 0~10bar suppose we calibrate
it at 0~4bar its mean its calibration range
is
0~4bar and instrument range 0~10bar
Is This Answer Correct ? | 26 Yes | 1 No |
Answer / raja
CAlibration range is the range upto which we have marked
readings on a guage ( for e.g) .
Instrument range is the range upto which instrument can
detect measurand .
calibration range is always less than Instrument range.
Dead zone and hysteresis may take position between the
difference in instrument and calibration ranges
Is This Answer Correct ? | 3 Yes | 0 No |
any body who gone thru recent iocl interview.INSTRUMENTATION questions. I need.
What is the advantage of using locally fabricated hook ups over Pre fabricated hook ups in process instrumentation?
0 Answers Engineers India Limited,
how can we calibate ph meter. how can we calibrat conductivimeter how can we calibrat massflowmeter & it s working.
What is the different between ESD & PLC.
Thermocouple is a transducer but rtd is not transducer why? Both taking heat as a input and giving mv
what is difference between OP AMP and differential amplifier?
my temperature transmitter in field showing 700 degree , whereas in dcs showing 459 degree, how to resolve this?
Explain the usage of the PLC system?
if field commnictor 475 does not communicate with transmitter what will you do ?
how can you set a zero and span of transmitter?
how to measure low temperature(<100 dec celsisus) with non condact sensor
what are level of automation & how these are distinguished ?