What is the b/W dead volume and dwell volume ?
Answers were Sorted based on User's Feedback
Answer / falguni patel
Dead volume: It comprises the volume of HPLC system between
the point of injection to the point of detection, excluding
the part of column that contains packing.
Dwell volume: It includes the volume between pump and
column. in low pressure gradient it may be 1 mL & in high
pressure gradient it is to less.
Is This Answer Correct ? | 14 Yes | 2 No |
Answer / aks
Dead Volume:Volume between injector to detector excluding
column volume (and also called as extra column volume)
Dwell Volume:Volume form the point of mixture to the column
entry (Also called as system delay volume)
Is This Answer Correct ? | 10 Yes | 0 No |
Do you have experience with organic extractions or tissue digestions?
How do you give training to a new chemist
How will u perform degradation study in detectors other thann PDA in HPLC Methodology
In C8,C18....etc have 4&5 etc... microns Columns is there like that how to separate Chiral columns as per Micron size?
if partition coefficient value was positive means, what it indicates ?
What is intact assay method development.....?
what is mean by ambient temperature?
1)why use benzophenon,uracil in hplc calibration. 2)why use sucrose in polarimeter calibration. 3)what is S/N ratio, noise and drift in hplc 4)how we can find lod and loq in method validation
If compound Pka is 5 then which buffer you will select ?
signifcance of detector lenearity in calibration f hplc
In chromatograpbic purity 90% is there, in assay 70% is there. How will you prove remaining 30%.