What is the b/W dead volume and dwell volume ?
Answers were Sorted based on User's Feedback
Answer / falguni patel
Dead volume: It comprises the volume of HPLC system between
the point of injection to the point of detection, excluding
the part of column that contains packing.
Dwell volume: It includes the volume between pump and
column. in low pressure gradient it may be 1 mL & in high
pressure gradient it is to less.
| Is This Answer Correct ? | 14 Yes | 2 No |
Answer / aks
Dead Volume:Volume between injector to detector excluding
column volume (and also called as extra column volume)
Dwell Volume:Volume form the point of mixture to the column
entry (Also called as system delay volume)
| Is This Answer Correct ? | 10 Yes | 0 No |
If vendor having more imp than monograph then how to proceed? and how to set spec?
what is peak purity and its formula
Suppose we prepared pH buffer solution in lab.and calibrate against NIST solution, and same shall going to used upto 30 days. is it possible to stable pH and apperance ? suggest ?
why require to add enzymes in capsules dissolution only? is it require to tablet also?
what is biorelivent dissolution media?
why salicylic acid not use for dissolution.
2 Answers Cipla, Indo Baijin Chemicals,
i want to know about headspace and interview question on headspace
1 Answers Srikem Laboratories,
What is difference between chemical Assay and Hplc assay?
UV and PDA detector, which have less signal to noise ratio?
What are the limits for uniformity dosage content for tablets and capsules
Why 1.2% KCl is used for limit of stray light? and why limit of stray light is required in UV calibration?
what is the principle of assay of sodium nitrite?