What is the b/W dead volume and dwell volume ?
Answers were Sorted based on User's Feedback
Answer / falguni patel
Dead volume: It comprises the volume of HPLC system between
the point of injection to the point of detection, excluding
the part of column that contains packing.
Dwell volume: It includes the volume between pump and
column. in low pressure gradient it may be 1 mL & in high
pressure gradient it is to less.
Is This Answer Correct ? | 14 Yes | 2 No |
Answer / aks
Dead Volume:Volume between injector to detector excluding
column volume (and also called as extra column volume)
Dwell Volume:Volume form the point of mixture to the column
entry (Also called as system delay volume)
Is This Answer Correct ? | 10 Yes | 0 No |
why wetlab is called as wetlab
Which is the highly polar and highly non polar column in HPLC?
What is zero water in TOC Analysis?
Why we are using a Kf Factor range 4 to 6
why disodium tartarate used in karl-fischer instrument ?.
What is column in chromatography?
In BP2013, Loperamid HCl monograph. Assay by titration with 0.1N sodium hydroxide using hydrocloric acid 0.01N and reading the volume added between the 2 points of inflexion. I have a question that if the diluent solvent is ethanol is certainly consumed a amount of volume of titrant, so this volume must be eliminated on the result calculation or not apart from first point which is subtracted above.
In case of dissolution test, if there is variation within 6 individual units however the batch is complying as per S1 criteria, is it necessary to do investigation for the same? What should be the criteria for such variation?
please explian me about area normalisation method,diluted standard method,impurity vs impurity and which one has to be selected in method development
If inhouse hplc related substance method is completly diffrent from Usp for finished proďuct with diffrent impurities then how require to prove method equivalecy?
How do you verify the primary,secondary and tertiary amines by IR-Spectroscopy.
why we are using k2cr207 in uv calibration ? why we r not using kmno4?
1 Answers Glochem, Sidmak Laboratories,