Why 1.2% KCl is used for limit of stray light? and why
limit of stray light is required in UV calibration?
Answers were Sorted based on User's Feedback
Answer / rk
First we have to know what is stray light. Stray light
defined as the light which is coming from the intended
source,not in intended path and also not with the intended
source, and KCl is used because it acts as stray filter
| Is This Answer Correct ? | 41 Yes | 9 No |
Answer / s dinesh
1.Stray light means the rays of light other than intended use.These make hindrance in the path of mono-chromatic light of use i.E.Why it is calculated. Its value must be greater than 2.0 at about 200nm.
2. 1.2 % soln is the conc of KCl which is capable restricting the undesired Radiations.
| Is This Answer Correct ? | 17 Yes | 0 No |
Answer / minhaz the great
Stray light is light other than intended source. 1.2% w/v solution of KCl in a1cm at 200nm gives 2.0 absorbance and if stray light is present it shows deflection in absorbance. Hence it can act as filter
| Is This Answer Correct ? | 8 Yes | 0 No |
Answer / vijay kumar y
1.0% NaI also used for stray light calibration at 210-260
nm..and it gives max. at 220.
in the same way
5.0% NaNo2 also used for stray light calibration at 300-385
nm..and it gives max. at 340.
| Is This Answer Correct ? | 11 Yes | 14 No |
why cone formation during dissolution?
Optical Rotation*0.9477*100*200] Assay Of dextrose= ________________________________ 99.8* 200 why Factor=0.9477
1 Answers Claris Lifesciences, Inven Pharmaceuticals, Levo,
Why we are use 0.005M Sulphuric acid, in control of absorbance performance test UV Spectrophotometer calibration
what is difference b/w spectroscopy, Spectrometry and Spectrophotometry ?
which procedure we follow if 6th month of accelerated sample fail?
on what basis SOP's can be changed? IF mobile phase ratio is 70:30 to how much % can we change the ratio in HPLC process
what is Hilic chromatography?
Why only lead standard solution is uesd to perform heavy metal test.
why base line not starts from zero in gc?
1 Answers Aurobindo, FDC, Reddy Labs,
While performing TOC sst analysis Zero shift disabled & sample analysis zero shift enabled why?
acceptance criteria for lod & loq by standard deviation of response and slope??
WHAT IS DIFFERENCE B/W MALTINGRANG/DSC