How do we fix the sample concentaryion in hplc method
development. what is the basis?Thanks in advance
Answers were Sorted based on User's Feedback
Answer / vipul
For Assay:- Sample concentration should be lower than higher
detector linearity concentration.
For Related substances:- Sample concentration should be
decided on the basis of LOD and LOQ determination
e.g sample concentration 1000ppm
Standard low load concentration 5ppm(For specified imp
limit-0.5%)
LOQ-should be between 0.05% To 0.09% of sample concentration.
| Is This Answer Correct ? | 8 Yes | 2 No |
Answer / pramod
It is also depend upon peak response.
How ever we have to select sample concentration in rs on the basis of LOQ
"Loq should be less than Reporting threshould"
Accordign to that we have to select sample Concentration
| Is This Answer Correct ? | 2 Yes | 0 No |
Answer / rajareddy
based on the peak response ,peak height and LOD and LOQ Signal. Ex:sample concentration 1mg/ml than LOQ value (0.1%mg/ml) LOD value(0.03mg/ml)
| Is This Answer Correct ? | 0 Yes | 0 No |
Answer / asha latha
IF ALL IMPURTITIES AT SPEC LEVEL ARE MEETING S/N RATIO MORE
THAN 60 AT THAT PARTICULAR CONCENTRATION AS WEL AS NO
COLUMN OVERLOAD.
| Is This Answer Correct ? | 0 Yes | 6 No |
In the test for control of absorbance in UV calibration why do we use only potassium dichromate and what is purpose of taking a specified amount of 57.0-63.0mg?
8 Answers Aurobindo, Cipla, FDC,
why we cannot analysis below 200nm in u.v?
defination of wavelength & wave number
What is control room temperature and which guide line says?
acetic acid is weak acid & perchloric acid is strong acid.but how perchloric acid is soluble in acetic acid
WHAT IS THE PURPOSE WE USE HCL, CHOPPER, BURNER, MONOCHROMATOR in AAS? PLEASE EXPLAIN BRIEFLY
IF WE INCREASE FLOW RATE RT DECRESE.WHY
Why we use methanol as solvent in kf standardisation. Why not other solvent?
2 Answers Caplin Point, Epione Labs, Hetero,
How to standardization of 1N H2SO4?
WHAT IS USE OF ION PAIR REAGENTS?
Many times I don't got a caffeine peak in calibration of hplc using guard column ❓
In capadity factor formula is K'= (RT of main compound/RT of first eluted peak)-1 in this formula what is 1