IN pH CALIBRATION,CALIBRATION IS DONE FOR pH METER OR
ELECTRODE?HOW WILL YOU JUDGE THAT THE METER TO BE
CALIBRATED WITH 4,7,9.2 BUFFERS?EXPLAIN TEMPERATURE EFFECTS
ON pH.
Answer Posted / prabhu
Hi dear the calibration is also one of the system
suitability test,whether the system working satisfactory or
not physically so when you do the calibration for pH meter
with standard pH buffer solution (which is from NIST-it may
accurate solution for actual pH)and once if your instrument
has calibrated,you can cross verify your insrument against
those pH buffer solution (NIST) and the acceptable range is
+/- 0.02 then you can conclude for ?
once if temperature raises the pH will decrease and its not
all cases
| Is This Answer Correct ? | 18 Yes | 5 No |
Post New Answer View All Answers
what you do when compound having dwel wavelength?
how to get accurate result for Residue on evaporation in purified water
give clarity of linearity and range in method validation
what is lod and loq ?,why use k2cr2o7 , kcl h2so4 in uv calibration ?,why use benzophenone & caffene acetone in hplc calibration ?,what is leading peak in hplc ?why we do the calibration of limit of stry light in hplc & uv ?
What is shaking level in GC?
how decide the clining method and cleaning method validation require for this perticular products?
all types of questions asked at the cipla interview
in monograph 7 imp are there but in vendor showing 5 imp do not req to include 2 imp for that what will be justification?
What is the requirement for brazil in the analytical method vslidation comparr to ich?
In HPLC calibration, caffeine is used as primary standard for wave length calibration due to caffeine is having dual maxima at 273 & 205 nm and one minima at 245 nm. Any body can give reference of these details from any pharmacopeia (with chapter no.) or any other authentic guideline?
how require to interprit the method precision data of hptlc?
how will you do the prep for unstable componds?
what is classification of elemental impurities? what is risk assement in elemental impurities?
what is the difference between potentiometric titration and karl fischer titration?
Why we used in n-butyl acetate water content terminology while in ethyl acetate we used moisture content terminology?