What is the right procedure for adjusting the gap voltage
of bently nevada vibration sensor?
Answers were Sorted based on User's Feedback
Answer / imran mansuri
for vibration sensor installation take digital multimeter
connect on proxymeter out signal + comma (-) now insert
first your proxyprobeon sahft untill it's not touch body
you will get 0 volt output now unscrew proxy untill you
get -9.0vdc or -10 vdc equal 50 mil this is procedure that
i know
Is This Answer Correct ? | 14 Yes | 2 No |
Answer / rana mujahid
first check all the physical checks like,prob material is
correct for service,cable length is proper accroding to
proximeter,termination,supply is proper(-24vdc),etc..
after this connect volt meter with proximeter b/w out and com.
and try to adjust exact -10vdc but -9.5 to -10.5 are ok just
configured in systm.
Is This Answer Correct ? | 7 Yes | 0 No |
Valve positioner and actuators in detail.
How to regulate the flow of Air in 4" line, with max flow of 330 nm3/hr and Pressure of 0.5 bar? please suggest the type of instrument required.
How we can re-range the DP transmitter, range is -0.5 to 2.0bara. new range is -0.5 to 5 bar
how to measure low temperature(<100 dec celsisus) with non condact sensor
Explain, and give a real-world example of MODULATING control
What are the different types of thermistors used in industries? How to the operation of typical thermistor?
what is remote mount with MTA option? (In case of Dual Sensor Vortex Flowmeter,rosemount has written in its catalog drawings).
in mass flow meter what is drive friquancy ?
How to calculate 47Liter H2 Gas cyilnder consumtion? Cylinder inlet Pressure is 200bar and Discharge pressure is 6.0bar. Flow is 100ml/min.For the above condition how many days one cylinder can be used?
what is formula for calculating flow if i have pressure and pipe diameter?
What is the purpose of MATLAB?
PID controller is used for which type of compensation…… a) lead comp. b) lag com. c) lead lag com.
12 Answers Engineers India Limited,