differance between L1 and L7 columns
Answers were Sorted based on User's Feedback
Answer / saravanan thangavel
L1 columns coated with octadecyl silane bonded to porous
silica, 3 to 10 µm
where L7 columns coated with ocylsiliane bonded to totally
porous microsilica particles 3 to 10 µm
| Is This Answer Correct ? | 18 Yes | 1 No |
Answer / satheesh kumar
USP designed LC Columns as L,L1 denotes-C18 Column,L7
denotes as C8
| Is This Answer Correct ? | 12 Yes | 0 No |
Answer / harsh
L1 columns coated with octadecyl silane bonded to porous
silica, 3 to 10 µm
where L7 columns coated with octylsiliane bonded to totally
porous microsilica particles 3 to 10 µm
l1 gives better resolution then l7.
l1 have more carbon loading so good for resolution
| Is This Answer Correct ? | 9 Yes | 1 No |
Answer / priyanka.r
According to usp code L1=C18 COLUMN
L7=C8 COLUMN
| Is This Answer Correct ? | 8 Yes | 3 No |
What are usp limits for total impurities,unknown impurity
If compound Pka is 5 then which buffer you will select ?
How to prepare 50 ppm ammonia solution form 25% ammonia solution?
What is linearity in HPLC and how will you test?
what is analytical chemistry all about?
For standardization of volumetric solutions, acceptance limit for concentration is considered as 10% of nominal concentration and %RSD as NMT 0.2% (for appropriate no. of replicates). But , if %RSD has not met, what should be further action?
How to choose concentration of LOQ solutions for RS analysis? Example.
1 Answers Abbott, Indoco Remedies,
what is the difference between Hydrochloric Acid & Hydrogen chloride
For limit test of heavy metals in BP, Method C require that the substance is ignited at a temperature not exceeding 800 °C. Why confines such the temperature?
What is diffrence between extractable volume and deliverable volume? Answer pls
Why dissolution test apparatus calibration with salycilic acid tablets was stopped by FDA&USP from the year of 2009?
What is similarity factor?