Answer / francisca
split injector is used for more concentrated solution whereas splitless injector is used for dilute solution.
Is This Answer Correct ? | 8 Yes | 2 No |
Answer / k.vineel kumar
Split liner is that in which it takes only i part from the
injection volume and dicards the remaining 4 parts through
vent outside if the split ratio is 1:5
Splitless is that in which if 1µl is the injection volume
it takes full of it and vent is not present in splitless
liner.
So ultimately the responses will be high in Splitless than
in split
Is This Answer Correct ? | 3 Yes | 2 No |
Answer / srikanth
Split injection: In this mode only a portion of the inject
sample (1/50 part)reaches the column and rest is vented to
waste. it is used for concentrated samples (> 0.1 mg/mL).
Splitless injection: In this mode all the sample injected
passes through the column. it is used for trace samples (<
0.1 mg/mL)
Is This Answer Correct ? | 2 Yes | 1 No |
Answer / gopi
the use of split is its increases the life time of column
and gc also it takes small amount of sample
Is This Answer Correct ? | 0 Yes | 1 No |
if the purity is 99.9%,is need to do assay?
What is colimn selection Criteria in HPLC analysis?
some compound havving uv region but not give much more response during hplc analysis
What is the difference between Normal phase method & Reverse method in HPLC ? Define application of both method.
My question about gas chromatography sulfur chemiluminsecence detector. I test unknown sample gas by GC-SCD (calibrated ) and the result of *H2S is 279 PPM , *but when I test the same sample with the GC-TCD (calibrated ) the value of *H2S is 2500 PPM . I'd like to inform you that both GCs are calibrated and have very good operation conditions with stable parameters . the question is if the sample gas with higher H2S over detection limits of SCD detector (1000 ppm). why the result it 279 ppm Best regards
How to do regeneration of Metacarb Pb plus column?
loss on drying we applied vaccum for some compounds and some compounds are at 105c why
How is the Flame test performed?
Why Deuterium lamp is used as UV light Source for UV.
from compund pka value how to select buffer and column?
2 Answers Emcure, Zydus Cadila,
As per USP what is RSD limit of volumteric solution of three replicate analysis
significane of detector wavelength accrucy in calibration of hplc